LC & CE Drivers A.02.17 or above Instrument Control Framework (ICF) A.02.04 or above Instant Pilot (G4208A) with firmware B.02.22 or above Lab Advisor B.02.10 or above
For details about supported software versions refer to the compatibility matrix of your version of the LC and CE Drivers
Communications
Controller-area network (CAN), Local Area Network (LAN) ERI: ready, start, stop and shut-down signals
Maintenance and safety-related features
Extensive diagnostics, error detection and display with Agilent Lab Advisor software Leak detection, safe leak handling, leak output signal for shutdown of pumping system, and low voltages in major maintenance areas
GLP features
Early maintenance feedback (EMF) for continuous tracking of instrument usage with user-settable limits and feedback messages. Electronic records of maintenance and errors.
Housing
All materials recyclable.
Table 3 Physical Specification of the Sample Cooler
Type
Specification
Comment
Weight
< 6 kg (< 13.2 lbs)
Dimensions (height × width × depth)
205 x 340 x 370 mm (8.1 x 13.4 x 14.6 inches)
Refrigerant gas
HFC-134a (0.042 kg)
Ozone depletion potential (ODP) = 0
Supply voltage
24 VDC
Current
10 A max.
Ambient operating temperature
4 – 40 ° C (39 – 104 ° F)
Ambient non-operating temperature
-40 – 70 ° C (-40 – 158 ° F)
Humidity
< 95 % r.h. at 40 °C (104 °F)
Non-condensing
Operating altitude
Up to 3000 m (9842 ft)
Safety standards:
IEC, EN, CSA, UL
Installation category II, Pollution degree 2
For indoor use only
ISM Classification
ISM Group 1 Class B
According to CISPR 11
Table 4 Performance Specifications of the Sample Cooler
Type
Specifications
Operating principle
High performance, low-energy consumption micro-compressor based cooler with ozone-friendly HFC-134a coolant (42 g), user-upgradable.
Temperature range
from 4 °C to 5 °C below ambient
Temperature settable
from 4 – 40 °C in 1 ° increments
Temperature accuracy (<25 °C, <50 % r.H.)
2 °C to 6 °C at a setpoint of 4 °C
The Agilent Infinity II Sample Cooler is not available for trade sales anymore and has been replaced by the Agilent InfinityLab Sample Thermostat.
Table 5 Physical Specifications of the Sample Thermostat (G7167-60101 and G7167-60201)
Type
Specification
Comments
Weight
<6 kg (< 13.2 lbs)
Dimensions (height x width x depth)
205 x 340 x 370 mm (8.1 x 13.4 x 14.6 inches)
Refrigerant gas
R600a (max. 0.030 kg)
Ozone depletion potential (ODP) =0 Global warming potential (GWP) =3
Supply voltage
24VDC
Current
10 A max.
Ambient operating temperature
4 °C to 40 °C (39 °F to 104 °F)
For sample cooling, ambient temperature ≥10 °C
Ambient non-operating temperature
-40 °C to +70 °C (-40 °F to +158 °F)
Humidity
< 95 % r.h. at 40 °C (104 °F)
Non-condensing
Operating altitude
Up to 3000 m (9842 ft)
Safety standards: IEC, EN, CSA, UL
Pollution degree 2
For indoor use only
ISM Classification
ISM Group 1 Class B
According to CISPR 11
Table 25 Initialization Mode Switches
SW1
SW2
SW3
SW4
SW5
SW6
Init Mode
0
0
0
0
0
0
Use Default IP Address
0
0
0
0
1
0
Use Stored IP Address
0
0
0
1
0
0
Use DHCP
Note: The setting ‘0’ (down) is essential.
Engineering Facts
12 facts · every claim linked to a page
Installation
The module requires additional clearance on the sides and rear for air circulation and electrical connections.
Ensures adequate ventilation and access for cables during bench installation.
Source
“It needs an additional 2.5 cm (1.0 inches) of space on either side and approximately 8 cm (3.1 inches) in the rear for air circulation and electric connections.”
Installation
The module must be operated horizontally, especially with a Sample Cooler or Thermostat, and leveled using a bubble level.
Prevents mechanical issues and ensures proper operation of the sampler and cooler/thermostat.
Source
“The module should be operated in a horizontal position, especially if a Sample Cooler or Sample Thermostat is installed. Use a bubble level to check the leveling of the sampler.”
Safety
The module is partially energized even when switched off if the power cord is plugged in.
Highlights electrical shock hazard during maintenance or service.
Source
“The module is partially energized when switched off, as long as the power cord is plugged in.”
Installation
The power connector must be easily reachable and unpluggable in case of emergency.
Ensures compliance with safety requirements for emergency disconnection.
Source
“Make sure the power connector of the instrument can be easily reached and unplugged.”
Safety
Only the power cord shipped by Agilent should be used with the instrument.
Prevents personal injury or equipment damage from incompatible power cords.
Source
“Never use a power cord other than the one that Agilent shipped with this instrument.”
Operation
For injection volumes below 2 µL, degassed mobile phases are required to avoid gas bubbles affecting precision.
Critical for achieving accurate low-volume injections.
Source
“Use degassed mobile phases for best injection-volume precision with injection volumes < 2 µL.”
Operation
The default draw speed is 100 µL/min, but lower speeds are recommended for viscous samples to avoid air bubbles.
Improves injection volume precision when handling viscous samples.
Source
“The default draw speed is 100 µL/min. This speed is suitable for the majority of applications, however, when using viscous samples, set the draw speed to lower speed for optimum results.”
Operation
The default eject speed is 400 µL/min; higher speeds shorten the injection cycle but should be avoided for viscous samples.
Optimizes cycle time while avoiding issues with viscous samples.
Source
“The default eject speed setting is 400 µL/min. When using large injection volumes, setting the eject speed to a higher value speeds up the injection cycle by shortening the time the metering unit requires to eject solvent at the beginning of the injection cycle (when the piston returns to the home p”
Maintenance
The peristaltic pump cartridge is a replaceable unit; the tubing inside is not replaceable.
Informs maintenance planning and spare parts ordering.
Source
“The peristaltic pump cartridge is a replaceable unit. The tubing inside the pump is not replaceable.”
Maintenance
When replacing the rotor seal, do not over-tighten the stator head screws; sealing force is automatic.
Prevents damage to the valve during maintenance.
Source
“Do not over-tighten the screws. The screws hold the assembly together and do not affect the sealing force. The sealing force is automatically set as the screws close the stator head against the valve body.”
Installation
Transporting the sampler with a Sample Cooler/Thermostat installed is only allowed for short distances; for longer distances, separate the units.
Avoids mechanical damage during relocation.
Source
“Transporting the sampler with a Sample Cooler/Sample Thermostat installed is only allowed for short distances. For longer distances, you must separate the units and send them independently.”
Safety
The leak sensor tubing is made of PVDF, which is incompatible with DMF.
Prevents solvent compatibility issues that could damage the leak sensor.
Source
“The tubing of the leak sensor is made of PVDF (polyvinylidene fluoride), which is incompatible with the solvent DMF (dimethylformamide).”