Contents

  1. Overview pp. 6-17
    Product descriptions of system modules and detectors
  2. Procedure pp. 18-42
    Installing modules, stack configuration, network integration, capillaries, leak and waste handling
  3. Procedure pp. 43-50
    LAN configuration, instrument configuration, Lab Advisor, add-ons
  4. Procedure pp. 51-62
    Best practices, preparing a run, operating the system
  5. Quality p. 63
    InfinityLab LC performance verification
  6. Overview pp. 64-68
    Typical bio-inert LC applications and SEC analysis
  7. Ordering pp. 69-75
    HPLC tool kit, heater devices, uptime kit, valve kits
  8. p. 27 table, 4 rows
  9. p. 38 table, 24 rows
  10. p. 75 table, 2 rows

Bio-inert capillaries

p/nDescription
5500-1264Capillary Ti 0.17 mm x 500 mm, SL/SLV Pump to multisampler
G5667-81005Capillary PK/ST 0.17 mm x 500 mm, RLO/RLO (Bio-inert) Multisampler to MCT
G5615-68755Capillary Kit Flow Cells BIO includes Capillary PK 0.18 mm x 1.5 m and PEEK Fittings 10/PK (p/n 5063-6591) MCT to DAD
G5664-68712Analytical tubing kit 0.25 mm i.d. PTFE-ESD DAD to fraction collector

Table 2: Compatibility of drain connectors and modules

Drain Connector TypeCompatible ModuleCompatible Module Type
DoubleG7116A/BColumn Compartment
SingleG7114A/BDetector
G7115A
G7117A/B/C
G7121A/B
G7123B
G7162A/B
G7165A
G7129A/B/CSampler
G7167A/B/C
G5668A
G7137A/B
G7157A
G4767A
G7122ADegasser
G7104A/CPump
G7110B
G7111A/B
G7112B
G7120A
G7131A/C
G7132A
G5654A
G4782A

Table 8: G4235A Solvent Selection Valve Head, Bio-inert - Stand Alone Valve

DescriptionAmountUnitComment
5067-4159 (12-position/13-port selector valve head, 210 bar, bio-inert)1ea
5067-4601 (Solvent selection tubing kit, 4 solvents)up to 4eaoptional

Engineering Facts

13 facts · every claim linked to a page
Selection

The pump's titanium-based flow path tolerates high salt and extreme pH, so it suits biological and high-salt methods.

Determines whether the system can run your salt/pH method without corrosion damage.

Source
“The titanium-based pump offers highest corrosion resistance against high salt concentration (2 M) and offers a handling of a wide pH range (1 – 13 , short term 14 ).”
Selection

The pump reaches 600 bar and 10 mL/min at 200 bar, enabling conventional, sub-2 µm and superficially porous columns.

Sets the column particle size and flow rate you can use.

Source
“It features a pressure range of up to 600 bar and a flow rate up to 10 mL/min (at 200 bar).”
Selection

Standard PEEK capillaries are pressure-sensitive, so bio-inert systems use stainless-steel cladded PEEK rated to at least 600 bar.

Prevents selecting plain PEEK tubing that would fail at system pressure.

Source
“Agilent uses stainless-steel cladded PEEK capillaries (PK/ST), which keep the flow path free of metal as required for bio-inert applications and ensure pressure stability to at least 600 bar.”
Installation

UHP-FF fittings must be tightened to a maximum torque of 0.8 Nm using the fitting tool or a 5 mm hex wrench.

Over-torque damages the fitting and capillary during assembly.

Source
“Use 5043-0915 (Fitting mounting tool) or a 5 mm hex wrench for fixing the fitting (maximum torque 0.8 Nm).”
Maintenance

Fittings must not be removed from used bio-inert capillaries because the expanded PEEK front end can be ripped off.

Removing fittings destroys the capillary and forces replacement.

Source
“Do not remove fittings from used capillaries.”
Installation

The bio-inert ZDV union has two different ends and capillaries must be installed in the marked sequence to avoid leaks.

Wrong assembly order damages the union insert and causes leaks.

Source
“The 5067-4741 (ZDV union (Bio-inert)) has two different connectors where capillaries need to be installed in the correct sequence.”
Safety

The solvent cabinet holds a maximum of 8 L total and no single bottle may exceed 2 L.

Defines safe solvent storage limits for the bench setup.

Source
“The solvent cabinet is designed to store a maximum volume of 8 L solvent. The maximum volume for an individual bottle stored in the solvent cabinet should not exceed 2 L.”
Safety

The leak sensor in the leak plane stops the running system once the leak detection level is reached.

Explains the automatic shutdown behavior relied on for safe operation.

Source
“The leak sensor in the leak plane stops the running system as soon as the leak detection level is reached.”
Installation

Drain connectors must be fitted on every module except the lowest one, where a waste tube is connected instead.

Ensures low-flow high-viscosity leaks drain properly through the stack.

Source
“Make sure that dripping adapters are correctly installed on each module in the LC stack, excluding lowest module.”
Installation

Stacking modules in the prescribed order optimizes the flow path for minimum delay volume.

Affects gradient delay and bench footprint at installation.

Source
“This configuration optimizes the flow path for minimum delay volume and minimizes the bench space required.”
Operation

The manual injector offers a standard 20 µL injection volume with optional loops from 5 µL to 5 mL.

Determines whether manual injection volumes match your method needs.

Source
“It offers a standard injection volume of 20 µL (optional 5 µL to 5 mL) and ensures highest injection accuracy.”
Operation

The detector covers 190–950 nm with a 120 Hz acquisition rate suited to fast LC.

Confirms detector compatibility with fast, high-throughput methods.

Source
“With its 120 Hz data acquisition rate the detector is perfectly suited for fast LC applications.”
Compliance

The instrument may only be operated by appropriately qualified and trained personnel.

Sets the personnel qualification requirement before commissioning.

Source
“The instrument may only be operated by appropriately qualified and trained personnel.”