Agilent 2100 Bioanalyzer Installation and Safety Guide
Technical Reference Summary
Covered Models & Document Scope
Covers the Agilent 2100 Bioanalyzer instrument in two versions: G2938C (electrophoretic assays and flow cytometry) and G2939A (electrophoretic assays only). Also covers the electrophoresis set G2947CA and flow cytometry set G2948CA for G2938C, plus spare parts and accessories.
Section-by-Section Technical Breakdown
The instrument power supply accepts any line voltage in the range 100-240 V ±10% at 50-60 Hz ±5%, so no voltage selector is present at the rear. Connecting the instrument to a line voltage higher than specified may damage it.
- Line voltage range: 100-240 V ±10%
- Line frequency: 50-60 Hz ±5%
- No voltage selector at the rear of the instrument
- Connect only to the specified line voltage
The instrument can be placed on almost any desk or laboratory bench. Approximately 8 cm (3.1 inches) of clear space is required at the rear for air circulation and electric connections.
- Approximately 8 cm (3.1 inches) clear space at the rear
- Required for air circulation and electric connections
The instrument operates within specified ambient temperature and humidity limits. Condensation from temperature fluctuations can damage the electronics, and the instrument should not be placed on or near vibrating surfaces. It is designed for a controlled electromagnetic environment (EN61326/A1) where RF transmitters such as mobile telephones should not be used in close proximity.
- Do not store, ship or use under conditions where temperature fluctuations could cause condensation
- If shipped in cold weather, leave in box and allow to warm slowly to room temperature
- Do not place on a vibrating surface or near vibrating objects
- Designed for controlled electromagnetic environment (EN61326/A1); avoid close proximity of RF transmitters such as mobile telephones
Table 1 lists the physical specifications including weight, dimensions, line voltage and frequency, power consumption, ambient operating and non-operating temperatures, humidity, operating and non-operating altitudes, and safety standards.
- Weight: 10 kg (22 lbs)
- Dimensions: 162 — 412 — 290 mm (width — depth — height)
- Line voltage: 100-240 VAC ±10%
- Line frequency: 50-60 Hz ±5%
- Power consumption: 30 W / 60 VA maximum
- Ambient operating temperature: 5-40 °C (41-104 °F)
- Humidity: <80% at 5-31 °C, decreasing linearly to 50% at 40 °C, non-condensing
- Operating altitude: up to 2000 m (6500 ft)
- Non-operating altitude: up to 4600 m (14950 ft)
- Safety standards: IEC, CSA, UL; Installation Category II, Pollution Degree 2
Installation requires placing the instrument upright on a suitable bench, removing the shipment lock, inserting the electrode or pressure cartridge into the lid and setting the chip selector accordingly, connecting the serial cable to the RS-232 port and PC, and connecting power. The heater plate and electrodes are sensitive to electrostatic discharge, so the operator must be grounded before inserting a chip.
- Place instrument upright on a bench meeting bench space requirements
- Open lid and pull metal lever on inside left to vertical position; remove shipment lock and store for future use
- Slide electrode or pressure cartridge into the lid and move lever to flat (closed) position
- Chip selector: position 1 for electrode cartridge, position 2 for pressure cartridge
- Connect serial cable to RS-232 port at rear and to PC serial port (COM-port)
- If PC has no serial port, use the provided USB/serial adapter cable
- Ensure power switch at rear is off before connecting power cord
- Power input socket accepts 100-240 VAC ±10%, 50-60 Hz; maximum power consumption 60 VA; no voltage selector
The power switch is located at the rear where the power cable plugs in. The status indicator lamp lights green if power is present and all self-tests pass; a red lamp indicates a hardware error. The instrument is ready to run samples once self-tests pass.
- Power switch located at rear of instrument where power cable plugs in
- Status indicator lamp lights green if power is present and all self-tests passed successfully
- Status indicator lamp lights red indicates a hardware error; refer to Troubleshooting section of the Maintenance and Troubleshooting Guide or Online Help
- Instrument ready to run samples after self-tests pass
When connecting via the USB/serial adapter cable, the driver must be installed on the PC. It is preinstalled on bundle PCs but must be installed manually on third-party PCs; the driver is on the 2100 Expert Software CD.
- Driver required on PC when using USB/serial adapter cable
- Preinstalled on bundle PCs; manual installation required on third-party PCs
- Driver found on the 2100 Expert Software CD
- Installation instructions in the readme file of the 2100 Expert Software
The software is pre-installed on bundle PCs but full functionality requires adding license key(s) supplied with the electrophoresis set and/or flow cytometry set. Instrument-PC communication is established by assigning a COM-port in the instrument context, after which serial number, vendor information and firmware version are displayed.
- Open Agilent 2100 Bioanalyzer expert software by double-clicking desktop icon
- Select Help > Registration and highlight the Add License tab
- Add the appropriate license key
- Assign a COM-port in the instrument context to establish instrument-PC communication
- Instrument-specific information (serial number, vendor information, firmware version) displayed when communication is established
Table 6 lists spare parts and accessories with reorder numbers, including RS232 cable, USB/serial adapter cable, fuse, gasket kit, adjustable clip, pressure adapter kit, vortex mixer adapter, electrode cartridge, chip priming station, test chip kits, and IKA vortexer.
- RS232-61601: RS 232 Cable (communication cable between PC and instrument)
- 8121-1013: USB/Serial Adapter Cable (connects RS232 cables to USB PC ports)
- 2110-0007: Fuse (fuse for power supply)
- G2938-68716: Gasket Kit (1 plastic adapter, 1 ring, 10 gaskets for chip priming station)
- 5042-1398: Adjustable Clip (used with a syringe to apply defined pressure for chip priming)
- 5065-4478: Pressure Adapter Kit (5 pressure adapters for pressure cartridge, G2938C only)
- 5065-9966: Vortex Mixer Adapter (replacement part for IKA MS2 Vortexer, 3 mounting screws)
- 5065-4413: Electrode Cartridge (no extra electrode pin set; pin-set not separately re-orderable)
- 5065-4401: Chip Priming Station (includes priming station, timer and 1 syringe clip)
- G2938-68300: Test Chip Kit for Electrophoresis Assays (autofocus chip, electrode/diode test chip and documentation)
- G2938-68200: Test Chip Kit for Flow Cytometry Assays (1 Cell Autofocus Chip and documentation, G2938C only)
- 5065-4428: IKA Vortexer MS3 type B (bundled to all Bioanalyzer systems but NOT available from Agilent standalone; order directly from IKA, part number 3617036)
Key Integration, Operating & Safety Requirements
| Specification / Condition | Engineering Implication | Applies To |
|---|---|---|
| Line voltage range is 100-240 VAC ±10% with line frequency 50-60 Hz ±5%; there is no voltage selector at the rear. p.16 ↗ | The instrument can be connected to any mains supply in this range without reconfiguration; connecting to a voltage higher than specified may damage the instrument. | G2938C and G2939A |
| Maximum power consumption is 30 W / 60 VA. p.20 ↗ | Power and cooling requirements for the installation site are modest; ensure the outlet and circuit can supply at least 60 VA. | G2938C and G2939A |
| Approximately 8 cm (3.1 inches) of clear space is required at the rear for air circulation and electric connections. p.17 ↗ | Bench layout must leave rear clearance to avoid overheating and to allow cable access. | G2938C and G2939A |
| Ambient operating temperature range is 5-40 °C (41-104 °F); assays are set up to be analyzed at a chip temperature of 30 °C, and if ambient temperature is above 30 °C the assays are analyzed using that temperature. p.20 ↗ | Laboratory temperature above 30 °C changes the effective assay temperature; temperature control of the lab is required for assays specified at 30 °C. | G2938C and G2939A |
| Humidity must be <80% at 5-31 °C, decreasing linearly to 50% at 40 °C, non-condensing. p.20 ↗ | High-humidity environments require dehumidification or air conditioning to prevent condensation damage to electronics. | G2938C and G2939A |
| Operating altitude is up to 2000 m (6500 ft); non-operating altitude is up to 4600 m (14950 ft). p.20 ↗ | Installation above 2000 m is outside the operating specification; storage and shipping up to 4600 m is permitted. | G2938C and G2939A |
| The instrument is a Safety Class I device with protective grounding; the power cord must be an Agilent Technologies cord designed for the region and the outlet must have a ground connection. p.7 ↗ | Never operate from an ungrounded outlet or with an unspecified power cord; use only Agilent-supplied cables to ensure safety and EMC compliance. | G2938C and G2939A |
| The heater plate and electrodes are sensitive to electrostatic discharge; the lid is grounded to leak off operator electrostatic energy. p.27 ↗ | The operator must be grounded before inserting a chip to prevent damage to the instrument. | G2938C and G2939A |
| Chip selector must be set to position 1 for the electrode cartridge and position 2 for the pressure cartridge. p.28 ↗ | Incorrect chip selector position will prevent correct operation for the installed cartridge type. | G2938C and G2939A |
| The status indicator lamp lights green if power is present and all instrument self-tests have passed; a red lamp indicates a hardware error. p.32 ↗ | A red status lamp requires referral to the Troubleshooting section of the Maintenance and Troubleshooting Guide or Online Help before running samples. | G2938C and G2939A |
| When starting the software, wait a few seconds until the self-test procedures are finished; otherwise the software may not recognize the instrument. p.31 ↗ | Premature software start can cause instrument communication failure; allow self-tests to complete first. | G2938C and G2939A |
| The USB/serial adapter driver is preinstalled on bundle PCs but must be installed manually on third-party PCs; the driver is on the 2100 Expert Software CD. p.33 ↗ | Third-party PC installations require driver installation before instrument-PC communication can be established. | G2938C and G2939A |
| Full software functionality requires license key(s) supplied with the 2100 Bioanalyzer electrophoresis set and/or flow cytometry set. p.34 ↗ | Without the appropriate license key, the software remains limited even though it is pre-installed. | G2938C and G2939A |
| The instrument is Laser Class 1 (IEC825-1, CFR1040.10); an interlock turns off the laser whenever the lid is not closed, and the lid cannot be completely opened and has a black light-absorbing surface. p.10 ↗ | System interlocks should never be disabled; no laser light is accessible during normal operation. | G2938C and G2939A |
| Sound pressure emission at the operator position is < 70 dB(A) in normal operation according to ISO 7779:1988/EN 27779/1991 (Type Test). p.19 ↗ | The instrument is suitable for normal laboratory acoustic environments without additional hearing protection. | G2938C and G2939A |
| The instrument is designed to operate in a controlled electromagnetic environment (EN61326/A1) where RF transmitters such as mobile telephones should not be used in close proximity. p.18 ↗ | Keep mobile phones and other RF transmitters away from the instrument to avoid electromagnetic interference. | G2938C and G2939A |
| If the instrument was shipped in cold weather, leave it in its box and allow it to warm up slowly to room temperature; let the instrument equilibrate to room temperature for one day to avoid condensation damage. p.18 ↗ | Cold-weather delivery requires a one-day equilibration period before power-on to prevent electronics damage from condensation. | G2938C and G2939A |
| The pressure adapter kit (5065-4478) and Test Chip Kit for Flow Cytometry Assays (G2938-68200) are for G2938C only. p.39 ↗ | Flow cytometry accessories cannot be used with the G2939A electrophoresis-only version. | G2938C |
Faults & Troubleshooting
| Code / Symptom | Possible Cause | Remedy | Note |
|---|---|---|---|
| Status indicator lamp lights red p.32 ↗ | Hardware error | Refer to the Troubleshooting section of the Agilent 2100 Bioanalyzer Maintenance and Troubleshooting Guide or the Online Help | |
| Software may not recognize the Agilent 2100 Bioanalyzer p.31 ↗ | Software started before the self-test procedures for the instrument are finished | Wait a few seconds until the self-test procedures are finished before starting the software |
Technical Data
| Parameter | Value | Condition |
|---|---|---|
| Weight | 10 kg (22 lbs) p.20 ↗ | |
| Dimensions | 162 — 412 — 290 mm (6.4 — 16.2 — 11.4 in) p.20 ↗ | Width — depth — height |
| Line voltage | 100-240 VAC ±10% VAC p.20 ↗ | Wide-ranging capability |
| Line frequency | 50-60 Hz ±5% Hz p.20 ↗ | |
| Power consumption | 30 W / 60 VA W / VA p.20 ↗ | Maximum |
| Ambient operating temperature | 5-40 °C (41-104 °F) p.20 ↗ | Assays set up to be analyzed on a chip temperature of 30 °C; if ambient temperat |
| Ambient non-operating temperature | -40-70 °C (-40-158 °F) p.20 ↗ | |
| Humidity | <80%, at 5-31 °C (41-89 °F), decreasing linearly to 50% at 40 °C (104 °F) % p.20 ↗ | Non-condensing |
| Operating altitude | Up to 2000 m (6500 ft) p.20 ↗ | |
| Non-operating altitude | Up to 4600 m (14950 ft) p.20 ↗ | For storage of the Agilent 2100 Bioanalyzer |
| Safety standards | IEC, CSA, UL p.20 ↗ | Installation Category II, Pollution Degree 2 |
| Sound pressure emission | < 70 dB (A) p.19 ↗ | At operator position, normal operation, according to ISO 7779:1988/EN 27779/1991 |