Setup, power connection, desiccant, software startup, and validation
Procedurepp. 23-34
Sample analysis, reference caps, cleaning, and ATR technique
Procedurepp. 35-38
Battery charging, desiccant recharge, and maintenance procedures
Orderingpp. 39-40
Spare parts list and ordering information
Specpp. 41-44
Technical specifications and safety information
pp. 45-52
Diamond ATR crystal, MAC address access, and library file transfer
p. 21table, 7 rows
p. 41table, 15 rows
p. 9table, 8 rows
p. 39table, 23 rows
p. 24table, 3 rows
p. 29table, 2 rows
Table 1 Diagnostic values
Value
Optimal (green)
Marginal (yellow)
Critical (red)
Comment
Energy (centerburst)
29,000 - 15,000
>29,000 or <15,000
>31,000 or <10,000
Indicates the overall alignment and proper gain adjustment for the system.
Battery life
AC or >30 min
5 - 30 min
<5 min
Source
1.9 A
>2.2 A or <1.6 A
>2.5 A or <1.0 A
Indicates a problem with the source control voltage or that the source has burnt out. Both voltage and amperage are displayed; however, amperage provides a sufficient diagnostic.
Laser signal
15,000 - 3,000
>15,000 or <3,000
>17,000 or <2,000
Can verify gross alignment errors even if the reference cap is not in place.
Detector temperature DTGS
35 - 44
<35 or >44
<30 or >48
Indicates a problem with the cooling circuitry or an ambient temperature over the specified range.
MCT
-51 to -49
<-51 or >-49
<-53 or >-47
The MCT detector is thermo-electrically cooled to approximately -50° C.
CPU temperature
10 - 75
>75
>80
Indicates the ambient temperature over the specified range.
Technical
Characteristic
Value
Interferometer geometry
High-throughput Michelson interferometer with fixed and moving flat mirrors
4300 HH FTIR Desiccant Cartridge (includes silica gel desiccant)
430-0164
USB cable 2 meter A to Micro B
G8180-68001
4300 HH FTIR D-ATR
G8180-68002
4300 HH FTIR Ge-ATR
G8180-68003
4300 HH FTIR Ext Ref
G8180-68004
4300 HH FTIR Diff Ref
G8180-68005
4300 HH FTIR GA Ref
G8180-20029
4300 Handheld FTIR wrist strap
250-0113
4300 Handheld FTIR stylus
For sampling accessories, upgrades, service contracts, repair and refurbishment services, contact Agilent or go to the Agilent website at www.agilent.com.
Table 2 Optimal uses of external reflectance sampling interfaces
Sampling Interface
Optimal Uses
External reflectance
• Suitable for analysis of smooth, opaque samples that reflect infrared light • Suitable for analysis of thin films and coatings greater than 1 micron in thickness on reflective metal surfaces
Grazing angle
• Suitable for analysis of ultra-thin films less than 1 micron in thickness • Suitable for analysis of very low-level contamination (nanometer thickness level)
Diffuse reflectance
• Suitable for analysis of samples that reflect little light
Table 3 Internal reflectance sampling interfaces
Sampling interface
Optimal Uses
Diamond ATR
• Ideal for analysis of solids, liquids, pastes and gel • Only the top 2-3 microns are analyzed
Germanium ATR
• Ideal for analysis of highly absorbing solids and liquids, e.g. carbon filled elastomers and rubbers • Only the top 0.5 to 2 micrometers is measured
See Appendix A for more details on the Diamond ATR crystals.
Engineering Facts
14 facts · every claim linked to a page
Safety
The instrument is not intrinsically safe, so it must only be used in atmospheres verified free of flammable materials.
Determines whether the analyzer can be deployed in hazardous or potentially explosive field environments.
Source
“4300 Handheld FTIR systems are NOT intrinsically safe. Use the system only in atmospheres that have been tested for flammable materials.”
Selection
The standard DTGS detector covers 4500 to 650 cm-1 and the instrument tolerates 0 to 50 C and up to 95% humidity.
Defines the spectral coverage and environmental envelope to match against the intended application.
Source
“The instrument comes with a DTGS detector capable of covering the spectral range of 4500 to 650 cm-1.”
Installation
After delivery the shipping case must be left at room temperature for several hours before opening to avoid condensation.
Prevents condensation damage to optics during initial setup and installation.
Source
“place the shipping case in a room-temperature environment and allow several hours for the contents of the container to reach the room ambient temperature.”
Installation
The instrument must be kept away from hot surfaces and sources of electromagnetic interference.
Guides placement of the instrument to avoid thermal and EMI-induced measurement errors.
Source
“Keep the instrument away from hot surfaces and any sources of electromagnetic interference.”
Operation
Batteries are hot-swappable, allowing continuous use without shutdown, and four batteries give about 8 hours of runtime.
Enables planning of field runtime and battery logistics without interrupting measurements.
Source
“You can open the front of the battery compartment and swap out a battery while the system is in use.”
Operation
Batteries cannot be charged inside the instrument and require the dedicated external charger.
Requires provisioning of the external charger for field battery rotation.
Source
“Batteries can only be charged in the appropriate battery charger. They do not charge in the 4300 Handheld FTIR system.”
Installation
Only the supplied 15 V DC power supply may be used, and a UL-approved surge protection strip is required.
Prevents damage from incompatible supplies and power surges during AC operation.
Source
“Use only the included 4300 Handheld FTIR power supply. Use of other power supplies may damage the 4300 Handheld FTIR.”
Installation
The power cord must be connected to the spectrometer first, then to the outlet and computer.
Defines correct power-up wiring sequence to avoid startup damage.
Source
“Connect the power cord to the spectrometer system first and then to the power outlet and computer to prevent any power startup damage or problems.”
Maintenance
The desiccant cartridge is recharged in a lab oven at about 120 C until the silica gel returns to dark blue.
Specifies the desiccant regeneration procedure needed to protect the optics from moisture.
Source
“The recommended temperature is approximately 120°C. Drying time may vary dependent on the amount of moisture.”
Operation
The instrument needs a 5-minute warm-up after power-on for reliable analysis, and 30 minutes before performance validation tests.
Sets the minimum stabilization time before collecting trustworthy data.
Source
“Allow the spectrometer to warm up for a period of 5 minutes for reliable analysis.”
Installation
The integrated CPU screen must not be rotated beyond vertical, or the flat flex cable will be damaged.
Prevents a common handling error that damages the display cable.
Source
“The integrated CPU screen rotated beyond the vertical position will result in damage to the flat flex cable connected to the screen.”
Installation
The USB cable must be inserted at a slight upward angle and the instrument placed on its side when connected.
Avoids damage to the USB port during PC communication and software updates.
Source
“The USB cable must be inserted at a slight upwards angle into the instrument USB connector”
Performance
Diagnostic values should be checked daily or weekly, while full performance validation runs quarterly or bi-annually.
Establishes the routine verification schedule to keep measurements within specification.
Source
“Performance validation should be run on a quarterly or bi-annual basis.”
Maintenance
The exterior may be wiped with a damp cloth and mild detergent but must never be fully immersed in water.
Defines safe cleaning limits to avoid water ingress into the handle and trigger.
Source
“Do not completely immerse the 4300 Handheld FTIR in water.”