Agilent 490-PRO Micro Gas Chromatograph User Manual
Technical Reference Summary
Covered Models & Document Scope
Covers the Agilent 490-PRO Micro Gas Chromatograph (Micro GC), a stand-alone process GC with one to four independent column channels, optional backflush, optional pressure regulators G3581-S0003 and G3581-S0004, optional LCD module, and PROstation software. Manual part number G3581-90006, fifth edition, June 2016.
Section-by-Section Technical Breakdown
The 490-PRO Micro GC can be equipped with one to four independent column channels, each a complete miniaturized GC with electronic carrier gas control, micro-machined injector, narrow-bore analytical column and micro thermal conductivity detector (µTCD). It is designed for stand-alone, unattended 24/7 operation in a process environment, with on-board data handling including peak integration, identification and calibration. Results are provided analog (4-20 mA), as digital Modbus TCP/IP, via serial communication to DCS/PLC/SCADA, or on an FTP server.
- One to four independent column channels
- Optional backflush protects stationary phase from moisture and CO2 and shortens analysis time
- On-board data handling: peak integration, peak identification, peak calibration
- Advanced sequence manager supports multistream analysis, auto calibration and calibration validation
- Built-in webserver provides instrument status and latest results
Carrier gas is connected at the back panel Carrier 1 or Carrier 2 port. The manual specifies pressure, purity and dryness requirements, recommends Gas Clean filters, and warns against plastic tubing. The instrument is configured for either He/H2 or N2/Ar carrier gas and the data system selection must match the physical connection.
- Carrier gas pressure: 550 kPa ± 10 % (80 psi ± 10 %)
- Purity: 99.999 % minimum, dry and free of particles
- Gas Clean filters recommended; flush filters and lines if not using nitrogen
- Do not use plastic tubing; use clean metal tubing
- Carrier gas configuration is either He and H2 or N2 and Ar
The Micro GC is built for analysis of gases and vapors only. Sample pressure should be between 0 and 100 kPa (0 to 15 psi) and sample temperature between 0 and 110 °C. The sample must be filtered, preferably through a 5-mm filter, and the external filter kit CP736729 is recommended.
- Sample pressure: 0 to 100 kPa (0 to 15 psi)
- Sample temperature: 0 to 110 °C
- Sample must be filtered, preferably through a 5-mm filter
- External filter kit CP736729 recommended
- Filter and dry sample before introduction if possible
The installation procedure covers pre-installation requirements, unpacking, packing list review, and seven steps: connect carrier gas, connect calibration gas or checkout sample, install power supply, connect to computer or local network, install PROstation, assign IP address, and complete Micro GC configuration in PROstation. It also covers restoring the factory default IP address, creating a test method, performing a series of runs, shut down procedure and long storage recovery procedure.
- Step 1: Connect carrier gas
- Step 2: Connect to calibration gas or checkout sample
- Step 3: Install power supply
- Step 4: Connect to computer or local network
- Step 5: Install PROstation
- Step 6: Assign IP address
- Step 7: Complete Micro GC configuration in PROstation
- Restore the factory default IP address procedure
- Create the test method and perform a series of runs
- Shut down procedure and long storage recovery procedure
This section describes carrier gas, Micro Electronic Gas Control (EGC), inert sample path, injector, column types (Molsieve 5Å, CP-Sil 5 CB, CP-Sil 13 and CP-Sil 19 CB, PoraPlot 10 m, Hayesep A 40 cm heated, COX and AL2O3/KCl, MES (NGA) and CP-WAX 52 CB), column conditioning, backflush option and TCD detector. It includes tuning the backflush time for standard and HayeSep A channels and how to disable backflush.
- Column types: Molsieve 5Å, CP-Sil 5 CB, CP-Sil 13 and CP-Sil 19 CB, PoraPlot 10 m, Hayesep A 40 cm heated, COX and AL2O3/KCl, MES (NGA) and CP-WAX 52 CB
- Column conditioning procedure
- Backflush option protects stationary phase and shortens analysis time
- Tuning backflush time except on a HayeSep A channel
- Tuning backflush time on a HayeSep A channel
- To disable backflush
This section covers the optional pressure regulators G3581-S0003 and G3581-S0004, including their installation and use.
- G3581-S0003 pressure regulator
- G3581-S0004 pressure regulator
The interfaces section covers access to connection ports, Ethernet connections (PROstation, firewall configuration, FTP server, Modbus TCP/IP, web server, multiple 490-PRO analyzers), USB WIFI (access via wireless network, access via PROstation, change USB AP name and passphrase), I/O connections (external digital I/O, external analog I/O), serial COM connections (connector layout, stream selector valve, Modbus serial), LCD screen, USB connection and interface examples.
- Ethernet RJ45 connector with LAN indicators: red LED transmit data, green LED receive data
- COM 1: RS-232 communication interface
- COM 2: RS-232 (2-wire) communication interface
- COM 3 and COM 4: RS232, RS422, or RS485 communication interface
- Digital I/O signals such as start_stop, ready_out, and start_in
- Analog I/O external analog signals
- SD card slot has no function supported
- Modbus TCP/IP and Modbus serial supported
- FTP server and web server supported
- USB WIFI access point with configurable SSID and passphrase
Covers installation requirements for the LCD module including environmental requirements, space requirements and power source, plus installation, inspection, unpacking, packing list, connecting the LCD module, setup, mechanical product specifications, connectors, shipping, cleaning and disposal instructions.
- Environmental requirements for LCD module
- Space requirements
- Power source requirements
- LCD module installation and connection
- How to setup the LCD
- Mechanical product specifications
Describes the 490-PRO cycle and stream selector, and the 490-PRO chromatographic run and electronic pressure control.
- 490-PRO cycle and stream selector
- 490-PRO chromatographic run and electronic pressure
Lists system requirements for hardware and software, and describes how to install PROstation.
- Hardware system requirements
- Software system requirements
- PROstation installation procedure
Covers the main menu, login procedure, main menu functions (configured instrument menu, control menu, virtual instrument), instrument configuration, 490-PRO configuration, communication frame (Ethernet communication, find instruments on the network, assign new IP address), configuration frame (hardware tab, user settings tab, PROstation tab, automation tab, info tab), exit configuration, services frame (calibrate pressure sensors, reboot 490-PRO Micro GC) and end configuration.
- Login procedure
- Find instruments on the network
- Assign new IP address
- Hardware tab, user settings tab, PROstation tab, automation tab, info tab
- Calibrate pressure sensors
- Reboot 490-PRO Micro GC
Covers login, instrument method setup, user roles (method developer/admin, service engineer/service, operator/read only mode), off line control and toolbar instrument control.
- User roles: method developer (admin), service engineer (service), operator (read only mode)
- Off line control
- Toolbar instrument control
Covers import/export functions (import FTP service file, import Galaxie ascii file, import OpenLAB CDS EZChrom ascii file, export as Galaxie ascii file, export as OpenLAB CDS EZChrom ascii file) and wizards (method wizard, application wizard, sequence wizard, Modbus wizard).
- Import FTP Service file
- Import Galaxie ascii file
- Import OpenLAB CDS EZChrom ascii file
- Export as Galaxie ascii file
- Export as OpenLAB CDS EZChrom ascii file
- Method Wizard, Application Wizard, Sequence Wizard, Modbus Wizard
Covers the view menu including application workspace and user workspace.
- Application workspace
- User workspace
Covers instrument setup (490-PRO Micro GC tab, instrument common tab, instrument method channel tab), integration events (chromatogram markers, set peak width, set threshold, set solvent threshold, estimate threshold, set minimum height/area, turn integration on/off, start/stop peak now, add peaks/grouping, split peak, baseline processing, shoulder peaks, set skim ratio, tangent skim next peaks on/off, tangent skim rear/front, exponential skim rear/front), peak identification (peak identification table, how to build an identification table, identification table columns, identification process, moving retention window), peak calibration (calibration parameters channel dependent and channel independent, use GOST calibration, initial calibration, prepare a calibration method, calibration chart) and method advanced (method properties).
- Instrument setup tabs: 490-PRO Micro GC, instrument common, instrument method channel
- Integration events and chromatogram markers
- Peak identification table and identification process
- Peak calibration parameters (channel dependent and channel independent)
- GOST calibration support
- Initial calibration and calibration chart
- Method properties
Covers application functions including normalize, synchronize, calorific power (energy calculation setup, ISO 6976 / GOST 31369, GPA2172 / ASTM-D3588, GPA 2172, ASTM D3588, GOST 22667, GOST 31369, accounting for water - partially saturated natural gas, sum C6+ unidentified components, component concentration by difference), verification check, alarms, analog outputs, timed relays, analog inputs, digital inputs, local user interface (LCD), system status and info parameters, channel specific status parameters, channel specific result parameters, component specific result parameters, energy meter result parameters, I/O parameters and API21 parameters.
- Calorific power calculations per ISO 6976 / GOST 31369, GPA2172 / ASTM-D3588, GPA 2172, ASTM D3588, GOST 22667, GOST 31369
- Accounting for water - partially saturated natural gas
- Sum C6+ unidentified components
- Component concentration by difference
- Verification check and alarms
- Analog outputs, timed relays, analog inputs, digital inputs
- System status and info parameters
- Channel specific status and result parameters
- Component specific result parameters
- Energy meter result parameters
- I/O parameters and API21 parameters
Covers automation sequence settings (auto start sequence on power-up, run sequence continuously, continuous analysis, times to repeat sequence, number of repeatings, run cycle time, home position, stream ahead scheduling), sequence table (sample type, replicates, calib. level, stream #, flush time, solution slot #), verification properties and table, calibration properties and table, site information, Modbus setup (process settings tab, synchronization with Modbus master, Modbus communication settings, serial communication settings, floating point type conversion, Int32 bit type conversion, shift Modbus addresses, registers setup tab, remote system synchronization, reading sample results, reading stream specific results, fixed values, execute commands, advanced Modbus information, Modbus bridge, Modbus parameter ID reference), FTP service, USB storage (setup, menus, settings, browse contents, test directly, save data during automated runs), real time clock and reprocess list.
- Auto start sequence on power-up
- Run sequence continuously
- Continuous analysis
- Times to repeat sequence and number of repeatings
- Run cycle time and home position
- Stream ahead scheduling
- Sequence table: sample type, replicates, calib. level, stream #, flush time, solution slot #
- Verification properties and table
- Calibration properties and table
- Modbus setup and parameter ID reference
- FTP service and USB storage
- Real time clock and reprocess list
Covers starting the analyzer, stopping column reconditioning, stop, upload, download, solutions, instrument status (automation, GC, GC channel, enhanced status), stream selector test, control reset I/O, control test I/O, reset alarms, reboot instrument and clear error log.
- Start the analyzer
- Stop column reconditioning
- Upload and download methods
- Solutions
- Instrument status: automation, GC, GC channel, enhanced status
- Stream selector test
- Reset I/O and test I/O
- Reset alarms, reboot instrument, clear error log
Covers integration report, application report, stream application report, diagnostics, print integration/application report and auto print application report after calibration or alarm.
- Integration report
- Application report
- Stream application report
- Diagnostics
- Print integration/application report
- Auto print application report after calibration or alarm
Covers chromatogram, calibration options (single level calibration, multilevel calibration, offline calibration, online calibration, Rw calibration, relative RF), setting up a typical single level calibration (environment, sequence), setting up a typical multilevel calibration (environment, calibration of the multilevel curve, sequence), single point calibration with multiple calibration mixtures, multiple point calibration with multiple calibration mixtures, and calibration validation (verification run, calibration limits).
- Single level and multilevel calibration
- Offline and online calibration
- Rw calibration and relative RF
- Single point calibration with multiple calibration mixtures
- Multiple point calibration with multiple calibration mixtures
- Calibration validation: verification run and calibration limits
Covers general setup, cases preparation, integration report, and four cases: analog output, alarms, timed relays and digital inputs.
- General setup
- Cases preparation
- Case 1: Analog Output
- Case 2: Alarms
- Case 3: Timed Relays
- Case 4: Digital Inputs
Describes error handling with four error classes (Class 0 internal event, Class 1 advisory fault, Class 2 critical error with error LED ON, Class 3 fatal error with instrument shutdown, error LED and buzzer ON) and provides the error list with error number, class, event/error code, description and action needed. Error codes are reported as CLNNN where C is error class, L is location (0=mainboard, 1-4=channel) and NNN is error number.
- Class 0: internal event, no influence on proper functioning
- Class 1: advisory fault, instrument continues
- Class 2: critical error for logging, error LED ON, run stopped
- Class 3: fatal error for logging, instrument shutdown, error LED and buzzer ON
- Error code format CLNNN: C=error class, L=location, NNN=error number
- Locations: 0=mainboard, 1=channel 1, 2=channel 2, 3=channel 3, 4=channel 4
- All Class 1 and higher errors logged in flash memory
Covers setting up the WinDCS communication including serial communication settings, Ethernet communication settings, general Modbus communication settings, WinDCS Modbus table (register type, data type) and WinDCS Modbus table setup.
- Serial communication settings
- Ethernet communication settings
- General Modbus communication settings
- WinDCS Modbus table: register type and data type
- WinDCS Modbus table setup
Covers operation of the history log application including starting the application, setup for data download, data download, setup for report, report control, chromatogram control, exit history log, and report data (header, calibration results, analysis data, avg/min/max, power on, alarm status change, parameter change).
- Starting the application
- Setup for data download and data download
- Setup for report and report control
- Chromatogram control
- Report data: header, calibration results, analysis data, avg/min/max, power on, alarm status change, parameter change
Key Integration, Operating & Safety Requirements
| Specification / Condition | Engineering Implication | Applies To |
|---|---|---|
| Carrier gas supply pressure must be 550 kPa ± 10 % (80 psi ± 10 %) with purity 99.999 % minimum, dry and free of particles. p.29 ↗ | Size the carrier gas regulator and plumbing to deliver this pressure; use Gas Clean filters to remove moisture and oxygen. | whole series |
| Sample pressure should be between 0 and 100 kPa (0 to 15 psi) and sample temperature between 0 and 110 °C. p.31 ↗ | Condition the sample to these limits before the injector; use the external filter kit CP736729 and a 5-mm filter. | whole series |
| The instrument is configured for either He and H2 or N2 and Ar carrier gas; the data system carrier gas selection must match the physical connection. p.30 ↗ | Changing the plumbed carrier gas type requires changing the corresponding carrier gas type in the data system. | whole series |
| Do not use plastic tubing for carrier gas because air diffuses through it, causing noisy baselines and decreased sensitivity; use clean metal tubing. p.29 ↗ | Specify flamed or chromatographically clean metal tubing for all carrier gas lines. | whole series |
| Error code format is CLNNN where C is error class (0=diagnostic, 1=advisory, 2=critical, 3=fatal), L is location (0=mainboard, 1-4=channel) and NNN is error number. p.719 ↗ | Decode the reported error code to identify the affected channel and severity before troubleshooting. | whole series |
| Class 2 critical errors light the error LED and stop the run in progress; Class 3 fatal errors light the error LED, sound the buzzer and shut down the instrument. p.718 ↗ | Treat Class 2 and Class 3 errors as requiring immediate corrective action; Class 1 advisory faults allow the instrument to continue. | whole series |
| Error 6 'Too low pressure' is triggered when pressure drops below 35 kPa; error 22 'Pressure too high' is triggered when pressure exceeds 450 kPa for at least 2 minutes. p.719 ↗ | Set carrier gas supply and pressure regulator to maintain pressure well within 35-450 kPa during operation. | whole series |
| Temperature faults 11, 12 and 13 are triggered when the sample line, injector or column temperature is not reached within 35 minutes (heater error). p.720 ↗ | Investigate heater or temperature sensor faults if temperature setpoints are not reached within 35 minutes. | whole series |
| Error 16 'Low power supply' is triggered when voltage is below 10 Volt. p.720 ↗ | Ensure the power supply delivers adequate voltage; recharge battery on portable Micro GC. | portable Micro GC |
| The instrument has one to four independent column channels, each with electronic carrier gas control, micro-machined injector, narrow-bore analytical column and micro thermal conductivity detector (µTCD). p.24 ↗ | Configure the number of channels according to the required analysis; each channel is a complete miniaturized GC. | whole series |
| The optional backflush protects the stationary column phase against moisture and carbon dioxide and shortens analysis time by preventing late-eluting compounds from entering the analytical column. p.24 ↗ | Enable backflush when analyzing samples with heavy compounds or moisture/CO2 to extend column life and reduce cycle time. | channels with backflush option |
| The 490-PRO Micro GC is designed to operate stand-alone, 24 hours a day, 7 days a week unattended and uninterrupted. p.24 ↗ | Design the installation for continuous process operation with remote monitoring via Modbus, FTP or web server. | whole series |
| Results can be provided analog (4-20 mA), as digital Modbus TCP/IP, via serial communication to a DCS, PLC or SCADA system, or on an FTP server. p.25 ↗ | Select the appropriate interface for integration with the plant control system; multiple interfaces are supported. | whole series |
| The mainboard provides Ethernet RJ45, COM 1 (RS-232), COM 2 (RS-232 2-wire), COM 3 and COM 4 (RS232/RS422/RS485), digital I/O, analog I/O and USB; the SD card slot has no function supported. p.28 ↗ | Plan cabling and connector access based on the available communication ports; do not rely on the SD card slot. | whole series |
| LAN indicators: red LED indicates transmit data, green LED indicates receive data. p.28 ↗ | Use the LAN LEDs for quick visual verification of Ethernet activity during commissioning. | whole series |
| The external filter unit must be hand-tightened into the female part followed by a 1/8 turn with a 7/16-inch wrench; orient the arrow on the female half towards the fingertight fitting. p.31 ↗ | Follow the specified tightening procedure to avoid leaks and ensure correct flow direction. | whole series |
| Hydrogen is combustible at concentrations from 4 % to 74.2 % by volume at atmospheric pressure and has the highest burning velocity of any gas. p.16 ↗ | Leak test all connections, lines and valves before operating with hydrogen; always turn off hydrogen supply at its source before working on the instrument. | whole series when using hydrogen carrier gas |
| Do not shut off column flow when the oven temperature is high, as this may damage the column. p.20 ↗ | Maintain carrier gas flow during high-temperature operation and follow proper shutdown procedures. | whole series |
| The instrument is properly grounded when shipped; no changes to electrical connections or chassis are needed for safe operation. p.20 ↗ | Do not modify grounding connections; use a power outlet with protective ground connection. | whole series |
| The 490-PRO Micro GC is designed for indoor use and has been tested and found to comply with the limits for a Class A digital device pursuant to part 15 of the FCC rules. p.18 ↗ | Install in a commercial/industrial environment; operation in a residential area may cause harmful interference requiring correction at user expense. | whole series |
Faults & Troubleshooting
| Code / Symptom | Possible Cause | Remedy | Note |
|---|---|---|---|
| Too low pressure p.719 ↗ | Pressure drops below 35 kPa | Check gas supply | Error 6, Class 1 |
| Pressure fault p.719 ↗ | Pressure state not ready after 5 minutes | Check gas supply or replace manifold | Error 7, Class 1 |
| Low battery 1 p.719 ↗ | Battery 1 low power (portable Micro GC only) | Recharge battery | Error 8, Class 1 |
| Low battery 2 p.719 ↗ | Battery 2 low power (portable Micro GC only) | Recharge battery | Error 9, Class 1 |
| Sample line sensor fault p.720 ↗ | Sample line temperature sensor error | Heater turned off | Error 10, Class 2 |
| Sample line temperature fault p.720 ↗ | Temperature not reached within 35 min (heater error) | Replace sample line heater | Error 11, Class 2 |
| Injector temperature fault p.720 ↗ | Temperature not reached within 35 min (heater error) | Replace module | Error 12, Class 2 |
| Column temperature fault p.720 ↗ | Temperature not reached within 35 min (heater error) | Replace module | Error 13, Class 2 |
| TCD Temperature limit activated p.720 ↗ | Hardware protection activated | Error 14, Class 1 | |
| EDS logging error p.720 ↗ | Unable to update EDS log | Call service | Error 15, Class 0 |
| Low power supply p.720 ↗ | Voltage < 10 Volt | Recharge battery | Error 16, Class 1 |
| Injector sensor fault p.720 ↗ | Injector temperature sensor error | Replace module | Error 17, Class 2 |
| Column temperature sensor fault p.720 ↗ | Column temperature sensor error | Replace module | Error 18, Class 2 |
| TCD control error p.720 ↗ | TCD voltage not or incorrectly set | Call service | Error 19, Class 2 |
| TCD calibration failed p.720 ↗ | Any error during TCD calibration | Replace module or TCD controller board | Error 20, Class 2 |
| Hardware reset p.720 ↗ | Instrument reset request from WS | Error 21, Class 2 | |
| Pressure too high p.720 ↗ | Pressure > 450 kPa for at least 2 minutes | Replace manifold | Error 22, Class 2 |
| Initialization error p.720 ↗ | During initialize | Call service | Error 23, Class 3 |
| Internal communication error p.720 ↗ | During/after initialization, between MPU and IOC/IOE | Call service | Error 24, Class 3 |
| Instrument EDS incorrect p.720 ↗ | Instrument Electronic Data sheet incorrect | Call service | Error 25, Class 3 |
| EDS incorrect p.720 ↗ | Electronic Data sheet incorrect | Call service | Error 26, Class 3 |
| Internal power failure p.720 ↗ | During/after initialization, internal supplies | Call service | Error 27, Class 3 |
| Temperature sensor disconnected p.722 ↗ | Temperature sensor not connect to instrument | Call Service | Error 53, Class 2 |
| Not ready to start run p.722 ↗ | Issued by Safety Control Object in Hardware domain. Bridge Call to GC domain (Reporting Not Ready To Start Run Error) | Check method | Error 54, Class 1 |
| Stream selection failed p.722 ↗ | Stream selector (VICI) failed switching | Check valve | Error 55, Class 1 |
| Ambient pressure or temperature alarm p.722 ↗ | Issued by Safety Control Object in Hardware domain whenever ambient temperature has passed a certain value. | Error 56, Class 1 | |
| Column cleaning p.722 ↗ | Instrument in column cleaning state | NA | Error 57, Class 1 |
| Equilibrating temperature zones p.722 ↗ | Instrument stabilizing after column cleaning | Wait until Ready | Error 58, Class 1 |
| IOC Communication error p.722 ↗ | MPU is not able to communicate with IOC | Call service | Error 76, Class 3 |
| Read mainboard EDS error p.722 ↗ | Not able to read Mainboard EDS | Call service | Error 77, Class 3 |
| Read channel controller EDS error p.722 ↗ | Unable to read EDS controller | Call service | Error 78, Class 3 |
| Read channel analytical module EDS error p.722 ↗ | Not able to read analytical module EDS | Call service | Error 79, Class 3 |
| Watchdog Error: Store Application report on flash error p.722 ↗ | Internal Software Error, cannot store application report on flash memory. | Auto reboot | Error 990, Class 3 |
| Watchdog Error: Store ErrorLog report on flash error p.722 ↗ | Internal Software Error, cannot store ErrorLog report on flash memory. | Auto reboot | Error 991, Class 3 |
| Watchdog Error: Instrument frozen (hazardous error) p.722 ↗ | Internal Software Error, software hanging | Auto reboot | Error 992, Class 3 |
| Watchdog Error: OOA Timer error p.722 ↗ | Internal Software Error, OOA Timer could not be created. | Auto reboot | Error 993, Class 3 |
| Watchdog Error: ACE reactor stopped p.722 ↗ | Internal Software Error, ACE reactor stopped. | Auto reboot | Error 994, Class 3 |
| Watchdog Error: Event pump stopped for 20 s p.722 ↗ | Internal Software Error, Event pump stopped. | Auto reboot | Error 995, Class 3 |
| Watchdog Error: IOC Fatal error 0 p.722 ↗ | Internal Software Error, IOC fatal error 0 | Auto reboot | Error 996, Class 3 |
| Watchdog Error: IOC Fatal error 1 p.722 ↗ | Internal Software Error, IOC fatal error 1 | Auto reboot | Error 997, Class 3 |
| Watchdog Error: IOC Fatal error 2 p.722 ↗ | Internal Software Error, IOC fatal error 2 | Auto reboot | Error 998, Class 3 |
| Watchdog Error: IOC Fatal error 3 p.723 ↗ | Internal Software Error, IOC fatal error 3 | Auto reboot | Error 999, Class 3 |
Technical Data
| Parameter | Value | Condition |
|---|---|---|
| Maximum operation altitude | 41 p.3 ↗ | |
| Carrier gas pressure | 550 kPa ± 10 % (80 psi ± 10 %) kPa p.29 ↗ | |
| Carrier gas purity | 99.999 % minimum % p.29 ↗ | |
| Sample pressure | 0 to 100 kPa p.31 ↗ | |
| Sample temperature | 0 to 110 °C p.31 ↗ | |
| Too low pressure threshold | 35 kPa p.719 ↗ | error 6 |
| Pressure too high threshold | 450 kPa p.720 ↗ | for at least 2 minutes, error 22 |
| Low power supply threshold | 10 Volt p.720 ↗ | error 16 |
| Temperature not reached timeout | 35 min p.720 ↗ | errors 11, 12, 13 |