Table 1 Gas Sensitivity Parameters
Gas
Gauge Constant S S = i+/i- x 1/P
Emission Setting for 0.01 Amp/Torr sensitivity (10 uA/micron)
Helium
1.5 per Torr
6.7 mA
Nitrogen
10 per Torr
1.0 mA
Air
10 per Torr
1.0 mA
where: S = Sensitivity • Torr; i+ = Ion current, Amps; i- = Grid (electron) current, Amps; P = Partial pressure of the gas, Torr
Table 8 Accessories and spare parts
Part Number
Description
L7430301
Glass ion gauge connector kit
R32443010
Cable, glass BA gauge standard, non-bakeable, XGS-600 gauge controller, 10 ft
R32443025
Cable, glass BA gauge standard, non-bakeable, XGS-600 gauge controller, 25 ft
R32443050
Cable, glass BA gauge standard, non-bakeable, XGS-600 gauge controller, 50 ft
Table 5 Degassing Ratings
Resistance Heating*
Electron Bombardment
Grid Voltage
6.3 VAC
Grid to Filament Voltage
+700 VDC
Grid Current
8.7 Amps
100 mA DC
Grid Temperature
1200 °C
1200 °C max
Filament Temperature
1400 °C
* All values in the Resistance Heating column are absolute maximum ratings. If these values are exceeded, sagging of the grid can occur. Electron Bombardment: 50 W absolute maximum
Table 6 NW25/NW40 KF and 2.75 CF flange Option
Flange
A
B
2.75 in DN 40 ConFlat flange
69.3 (2.73)
25.4 (1.00)
NW25 KF
40.1 (1.58)
25.4 (1.00)
NW40 KF
55.1 (2.17)
25.4 (1.00)
1.00 in tube
-
25.4 (1.00)
0.75 in tube
-
19.3 (0.75)
Engineering Facts
14 facts · every claim linked to a page
Selection
The gauge covers a wide pressure range from 1 mTorr down to 4x10-10 Torr, making it suitable for general-purpose vacuum measurement.
Defines the application envelope for selecting this gauge versus alternatives.
Source
“The 571 is useful as a general purpose vacuum gauge over the pressure range of 1 mTorr down to 4x10-10 Torr.”
Safety
The controller and vacuum system must be separately grounded to a common ground; a ground wire between chamber and controller chassis is unsafe.
Improper grounding can expose personnel to lethal voltages up to 1000 V.
Source
“Check that your Ion Gauge Controller and vacuum system are separately grounded to a common ground. Placement of a ground wire between the vacuum chamber and the controller chassis is not safe; large continuous currents could flow through it.”
Installation
The system ground must be verified to support at least 10 A before proceeding.
Ensures the grounding path can carry fault currents safely.
Source
“Before proceeding, test the system ground to be sure that it is complete and capable of supporting at least 10A.”
Safety
The gauge is not intrinsically safe and must not be turned on when combustible gases or mixtures are present.
Hot filament and high voltage can ignite combustible gas mixtures, causing injury and equipment damage.
Source
“As with all ionization gauges, this device is not intrinsically safe. Exercise extreme care when using this vacuum gauge while pumping or backfilling a system or in any other system condition which contains combustible gases or mixtures.”
Installation
Mount the gauge as close as possible to the measurement location through a large tubulation, preferably 1 inch OD, to avoid gauge pumping effects.
Low conductance causes the gauge to read lower than system pressure, especially below 10-8 Torr.
Source
“In general, the gauge tube should be mounted as close as possible to the desired location through a large tubulation (1" OD recommended).”
Operation
The tube can operate continuously in argon at 1x10-1 Torr without affecting filament life, but tube life in air may be shortened above 1x10-3 Torr.
Informs allowable operating pressure and gas-dependent lifetime limits.
Source
“When used in argon, the tube can operate continuously at 1x10-1 Torr with no effect on filament life. Tube life in air may be shortened above 1x10-3 Torr”
Maintenance
Do not bake the gauge at pressures above 10-5 Torr to avoid oxidation of gauge elements.
Prevents permanent degradation that would prevent reaching ultrahigh vacuum.
Source
“High temperature baking of the gauge should not be attempted at pressures above 10-5 to avoid oxidation of the gauge elements which can make it very difficult to attain ultrahigh vacuum pressure readings.”
Maintenance
Electron bombardment degassing applies up to 700 VDC to the grid connections, which can be lethal.
Requires appropriate precautions during degassing maintenance.
Source
“If electron bombardment degassing is employed, high voltages (up to 700 VDC maximum) will be applied to the grid connections; such high voltages can kill.”
Performance
The X-ray limit is approximately 2x10-10 Torr for nitrogen, and readings near this value should not be considered unequivocal.
Defines the lower useful measurement limit and uncertainty near it.
Source
“Therefore, as the typical value of X-ray limit is approached (2x10-10 Torr nitrogen equivalent), readings should not be considered unequivocal.”
Operation
The gauge is composition-sensitive; readings are expressed as equivalent pressure of one gas, usually nitrogen.
Requires gas correction factors when measuring gases other than nitrogen.
Source
“All ionization gauges are composition-sensitive and pressure readings derived from these gauges can only be expressed in terms of equivalent pressure of one gas, usually nitrogen.”
Maintenance
Do not use silicone oil or silicone grease on vacuum equipment; use powder-free butyl or polycarbonate gloves.
Prevents contamination that would degrade vacuum performance.
Source
“Do not use silicone oil or silicone grease. Use powder-free butyl or polycarbonate gloves to prevent skin oils from getting on vacuum surfaces.”
Maintenance
Do not clean aluminum parts with Alconox because it is incompatible with aluminum and causes damage.
Avoids corrosive damage to aluminum vacuum components.
Source
“Do not clean any aluminum parts with Alconox®. Alconox is not compatible with aluminum and will cause damage.”
Selection
The yttria-coated iridium filament operates at lower temperature than tungsten, resulting in less outgassing and resistance to accidental overpressure.
Reduces contamination and improves robustness for vacuum system selection.
Source
“Its yttria-coated Iridium Filament operates at a lower temperature than tungsten resulting in less outgassing. In addition, it will liberate gas during occasional accidental overpressure without immediate failure.”
Operation
Use a proper gauge controller that does not limit the broad range of the gauge or exceed maximum ratings.
Ensures full measurement range and prevents damage from incompatible controllers.
Source
“It is important to use a proper gauge controller that does not limit the broad range of the gauge, or exceed any of the maximum ratings noted in this manual.”