Screening and Confirmation of Anabolic Steroids Using Accurate Mass LC/MS
Technical Reference Summary
Covered Models & Document Scope
Application note describing an LC/MSD TOF (G1969A) method for screening and confirmation of anabolic steroids in urine; covers LC conditions, MS conditions, and method performance. No product model/size variants or ordering codes are given.
Technical Data
| Parameter | Value | Condition |
|---|---|---|
| Column | Agilent ZORBAX RRHT SB-C18 2.1 × 50 mm, 1.8 µm p.3 ↗ | |
| Column part number | 822700-902 p.3 ↗ | |
| Mobile phase A | 0.1% Formic acid/water p.3 ↗ | |
| Mobile phase B | Methanol p.3 ↗ | |
| Flow rate | 0.4 mL/min p.3 ↗ | |
| Column temperature | 55 °C p.3 ↗ | |
| Gradient | 55% B, hold 5 min; 55% to 75% B p.3 ↗ | |
| Analysis time | 14 min p.3 ↗ | |
| Vaporizer temperature | 450 °C p.3 ↗ | |
| Drying gas temperature | 350 °C p.3 ↗ | |
| Fragmentor voltage | 150 V p.3 ↗ | no collision-induced-ionization |
| Mass accuracy | 3 ppm p.4 ↗ | [M+H]+ ion, or less |
| Mass range | approximately 7000 m/z p.4 ↗ | |
| Photomultiplier voltage increase | 50 V p.5 ↗ | |
| Maximum counts | 1.2e5 counts p.5 ↗ | expected (M+H)+ |
| MRPL concentration | 2 ng/mL p.2 ↗ | or lower in urine |
| Reconstitution volume | 100 µL p.11 ↗ | |
| Injection volume | 4 µL p.11 ↗ | of the 100 µL reconstituted |
| Mass accuracy achieved | less than 2 ppm p.11 ↗ |
Operating Limits
- Minimum Required Performance Level (MRPL) concentration: 2 ng/mL or lower in urine p.2 ↗
- Mass accuracy requirement: 3 ppm or less p.4 ↗
- Mass accuracy achieved: less than 2 ppm p.11 ↗