Performance of the Agilent 7900 ICP-MS with UHMI for High Salt Matrix Analysis
Technical Reference Summary
Covered Models & Document Scope
Application note on the Agilent 7900 ICP-MS with UHMI (Ultra High Matrix Introduction) for high salt matrix analysis; covers direct measurement of samples up to percent-level TDS, with UHMI dilution factors up to 100; no product ordering variants.
Technical Data
| Parameter | Value | Condition |
|---|---|---|
| RF Power | 1600 W p.3 ↗ | 7900 ICP-MS with UHMI |
| Sampling Depth | 10 mm p.3 ↗ | 7900 ICP-MS with UHMI |
| Dilution Gas | 0.58 L/min p.3 ↗ | 7900 ICP-MS with UHMI |
| ORS Gas | 6 mL/min p.3 ↗ | H2 cell mode |
| Stabilization | 5 s p.3 ↗ | He cell mode |
| KED | 3.0 V p.3 ↗ | H2 cell mode |
| Spray chamber temperature | 2 °C p.3 ↗ | quartz spray chamber |
| Injector diameter | 2.5 mm p.3 ↗ | 7900 ICP-MS with UHMI |
| TDS limit (conventional ICP-MS) | 0.2 % p.2 ↗ | direct measurement, conventional ICP-MS |
| UHMI dilution factor | 100 x p.3 ↗ | UHMI dilution |
Operating Limits
- Maximum TDS for direct measurement by conventional ICP-MS: 0.2% (2000 ppm, 2 g/L) p.2 ↗
- Maximum salinity of hypersaline waters discussed: > 40% (Don Juan Pond) p.2 ↗
- UHMI dilution factor: up to 100x p.3 ↗