Induztr
Engineering Digest Document Viewer Related Models

Accelerating Modifier-Free SFC Separations with the 1260 Infinity II SFC System

Technical Reference Summary

Covered Models & Document Scope

Agilent Technical Overview describing an SFC separation method on the Agilent 1260 Infinity II SFC System (G4301A Control Module, G4782A Binary Pump, G4767A Multisampler, G7115A DAD with high-pressure SFC flow cell, G7116A Multicolumn Thermostat with G4237A four-column selection valve); column Agilent InfinityLab Poroshell 120 EC-C18, 3.0 × 150 mm, 2.7 µm (p/n 693975-302(T)); comparison of isocratic run, flow gradient and backpressure gradient.

Technical Data

Parameter Value Condition
BPR temperature 60 °C p.2 ↗
BPR pressure 150 bar p.2 ↗
Total run time 5.5 minutes p.2 ↗ no post time
Injection volume 4 µL p.3 ↗
Sample concentration 50 µg/mL p.3 ↗ nominal
Flow gradient rate 2.5 to 4 mL/min p.4 ↗ flow gradient
Column dimensions 3.0 × 150 mm p.2 ↗ Poroshell 120 EC-C18
Column particle size 2.7 µm p.2 ↗ Poroshell 120 EC-C18
Resolution (Rs) 1.69 p.5 ↗ backpressure gradient
Peak height increase 37.6 (+103 %) mAu p.5 ↗ backpressure gradient

Operating Limits

  • BPR pressure: 150 bar p.2 ↗
  • BPR temperature: 60 °C p.2 ↗
  • Minimum firmware for backpressure gradients (legacy modules): A/B/C/D 07.13 or higher p.2 ↗
  • Minimum driver for backpressure gradients: A.02.16 or higher p.2 ↗