Accelerating Modifier-Free SFC Separations with the 1260 Infinity II SFC System
Technical Reference Summary
Covered Models & Document Scope
Agilent Technical Overview describing an SFC separation method on the Agilent 1260 Infinity II SFC System (G4301A Control Module, G4782A Binary Pump, G4767A Multisampler, G7115A DAD with high-pressure SFC flow cell, G7116A Multicolumn Thermostat with G4237A four-column selection valve); column Agilent InfinityLab Poroshell 120 EC-C18, 3.0 × 150 mm, 2.7 µm (p/n 693975-302(T)); comparison of isocratic run, flow gradient and backpressure gradient.
Technical Data
| Parameter | Value | Condition |
|---|---|---|
| BPR temperature | 60 °C p.2 ↗ | |
| BPR pressure | 150 bar p.2 ↗ | |
| Total run time | 5.5 minutes p.2 ↗ | no post time |
| Injection volume | 4 µL p.3 ↗ | |
| Sample concentration | 50 µg/mL p.3 ↗ | nominal |
| Flow gradient rate | 2.5 to 4 mL/min p.4 ↗ | flow gradient |
| Column dimensions | 3.0 × 150 mm p.2 ↗ | Poroshell 120 EC-C18 |
| Column particle size | 2.7 µm p.2 ↗ | Poroshell 120 EC-C18 |
| Resolution (Rs) | 1.69 p.5 ↗ | backpressure gradient |
| Peak height increase | 37.6 (+103 %) mAu p.5 ↗ | backpressure gradient |
Operating Limits
- BPR pressure: 150 bar p.2 ↗
- BPR temperature: 60 °C p.2 ↗
- Minimum firmware for backpressure gradients (legacy modules): A/B/C/D 07.13 or higher p.2 ↗
- Minimum driver for backpressure gradients: A.02.16 or higher p.2 ↗