In-depth Peptide Mapping with Iterative MS/MS Acquisition on the Agilent 6545XT AdvanceBio LC/Q-TOF
Technical Reference Summary
Covered Models & Document Scope
Application note describing an LC/MS/MS peptide mapping workflow for CHO-cultured human IgG1 mAb using the Agilent AssayMAP Bravo, Agilent 1290 Infinity II LC, Agilent 6545XT AdvanceBio LC/Q-TOF with Iterative MS/MS, and MassHunter BioConfirm B.09.00; covers 15-minute and 30-minute gradients and 0.6 µg or 2 µg sample loading.
Technical Data
| Parameter | Value | Condition |
|---|---|---|
| Analytical column | AdvanceBio Peptide Mapping, 2.1 × 150 mm, 2.7 µm (p/n 653750-902) p.2 ↗ | |
| Column temperature | 60 °C p.2 ↗ | |
| Autosampler temperature | 4 °C p.2 ↗ | |
| Solvent A | 0.1 % formic acid in water p.2 ↗ | |
| Solvent B | 0.1 % formic acid in 90 % acetonitrile p.2 ↗ | |
| Flow rate | 0.4 mL/min p.2 ↗ | |
| Gas temperature | 325 °C p.2 ↗ | 6545XT AdvanceBio LC/Q-TOF |
| Sample loading amount | 0.6 µg p.2 ↗ | per analysis |
| MS1 mass tolerance | 10 ppm p.2 ↗ | BioConfirm search |
| MS2 mass tolerance | 30 ppm p.2 ↗ | BioConfirm search |
| Peptide length limit | 5 to 60 amino acids p.2 ↗ | BioConfirm search |
| False discovery rate | 0.1 % p.2 ↗ | peptide-spectrum match filter |
Operating Limits
- Column temperature: 60 °C p.2 ↗
- Autosampler temperature: 4 °C p.2 ↗
- Maximum missed cleavages: 2 p.2 ↗
- Peptide length range: 5 to 60 amino acids p.2 ↗
- False discovery rate: 0.1 % p.2 ↗