US EPA Method 524.2 with the Teledyne Tekmar Lumin P&T
Technical Reference Summary
Covered Models & Document Scope
Application note / study for Teledyne Tekmar Lumin purge-and-trap concentrator with AQUATek 100 autosampler and Agilent 7890B GC/5977A MS, demonstrating US EPA Method 524.2 VOC analysis; covers Lumin and AQUATek 100 operating conditions (Table 1) and GC/MS conditions (Table 2), not a product ordering datasheet.
Technical Data
| Parameter | Value | Condition |
|---|---|---|
| Valve oven temperature | 150 °C p.2 ↗ | Lumin, Table 1 |
| Transfer line temperature | 150 °C p.2 ↗ | Lumin, Table 1 |
| Sample mount temperature | 90 °C p.2 ↗ | Lumin, Table 1 |
| Purge ready temperature | 35 °C p.2 ↗ | Lumin, Table 1 |
| MCS purge temperature | 20 °C p.2 ↗ | Lumin, Table 1 |
| Standby flow | 10 mL/min p.2 ↗ | Lumin, Table 1 |
| Purge time | 11.00 min p.2 ↗ | Lumin, Table 1 |
| Purge flow | 40 mL/min p.2 ↗ | Lumin, Table 1 |
| Dry purge temperature | 20 °C p.2 ↗ | Lumin, Table 1 |
| Dry purge time | 1.00 min p.2 ↗ | Lumin, Table 1 |
| Dry purge flow | 100 mL/min p.2 ↗ | Lumin, Table 1 |
| Desorb time | 250 °C p.2 ↗ | Lumin, Table 1 (as printed) |
| Bake time | 2.00 min p.2 ↗ | Lumin, Table 1 |
| Bake temperature | 280 °C p.2 ↗ | Lumin, Table 1 |
| MCS bake temperature | 200 °C p.2 ↗ | Lumin, Table 1 |
| Bake flow | 200 mL/min p.2 ↗ | Lumin, Table 1 |
| Sample loop time | 0.35 min p.2 ↗ | AQUATek 100, Table 1 |
| Sample transfer time | 0.35 min p.2 ↗ | AQUATek 100, Table 1 |
| Rinse loop time | 0.30 min p.2 ↗ | AQUATek 100, Table 1 |
| Sweep needle time | 0.30 min p.2 ↗ | AQUATek 100, Table 1 |
| Presweep time | 0.25 min p.2 ↗ | AQUATek 100, Table 1 |
| Water temperature | 90 °C p.2 ↗ | AQUATek 100, Table 1 |
| Bake rinse drain cycles | 1 p.2 ↗ | AQUATek 100, Table 1 |
| Bake rinse drain time | 0.35 min p.2 ↗ | AQUATek 100, Table 1 |
| Calibration range | 0.5 to 50 ppb p.2 ↗ | all compounds except ketones and t-butanol |
Operating Limits
- Calibration range (ketones): 1.25 to 125 ppb p.2 ↗
- Calibration range (t-butanol): 2.5 to 250 ppb p.2 ↗
- Number of 0.5 ppb standards for MDL: 7 p.2 ↗