Maintaining Your GC — Agilent 8860 Gas Chromatograph

Technical Reference Summary

Covered Models & Document Scope

Maintenance procedures for the Agilent 8860 Gas Chromatograph (GC), covering capillary columns, Split/Splitless Inlet (SSL), Purged Packed Inlet (PPI), Packed Column Inlet (PCI), Cool On-Column Inlet (COC), Flame Ionization Detector (FID), Thermal Conductivity Detector (TCD), Electron Capture Detector (ECD), Nitrogen-Phosphorus Detector (NPD), Flame Photometric Detector Plus (FPD+), Auxiliary EPC module, Pneumatics Control Module (PCM), and valves. Manual part number G2790-90015, seventh edition, August 2024.

Section-by-Section Technical Breakdown

pp. 9-10
Overview of Maintenance

This section introduces the routine maintenance tasks for the Agilent 8860 GC and lists the components covered in the manual. It assumes basic knowledge of tool use and GC operation, including safely turning devices on/off, loading methods, changing temperatures/flows/pressures, making Swagelok connections, resetting service counters, and using the browser interface.

  • Covers maintenance for capillary columns, SSL, PPI, PCI, COC, FID, TCD, ECD, NPD, FPD+, Aux EPC, PCM, and valves.
  • Each component chapter includes consumables/parts list, exploded parts view, and detailed procedures.
  • Procedures assume user can safely turn devices on/off, load methods, change temperatures/flows/pressures, make Swagelok connections, reset service counters, and use the browser interface.
pp. 11-12
Tools and Materials Required for Maintenance

Table 1 lists common tools and materials needed for most GC maintenance procedures, with part numbers. Specific tools for each procedure are listed in step 1 of that procedure.

  • Wrench, angled, septum nut: 19251-00100
  • Wrench, open-end, 1/4-inch and 5/16-inch: 8710-0510 (included with GC ship kits)
  • Wrench, open-end, 9/16-inch and 7/16-inch: 8710-0803
  • Wrench, capillary inlet: G3452-20512 (included with GC ship kits)
  • Column cutter, wafer, 4/pk: 5181-8836 (included with GC ship kits)
  • T-20 Torx key or screwdriver: 8710-1807
  • T-10 Torx key or screwdriver: 8710-2140
  • 3-mm hex key wrench: 8710-2411
  • Electronic flow meter(s) or bubble meter(s) capable of calibrated measurements at 1, 10, and 100 mL/min flow ranges
  • Electronic leak detector
  • Magnifying loupe, 20X: 430-1020
  • ESD wrist strap: 9300-1408
  • Cleaning brushes—FID cleaning kit: 9301-0985
  • Cleaning brushes—for split/splitless inlet split vent fitting, FID and collectors: 8710-1346
  • Jet cleaning wire (.010-inch)
  • Gloves, clean, lint-free, nylon: large 8650-0030, small 8650-0029
  • Steel wool, 0- or 00-grade (for cleaning an inlet’s septum seating surfaces)
pp. 13
Preparing for Maintenance

Before routine maintenance, the GC must be made ready by setting low temperatures, reducing flows, turning off and unplugging the GC, venting an MSD, and making other settings to prevent damage. A general standby state can be achieved via the browser interface using Maintenance > Instrument > Perform Maintenance > Maintenance Mode > Start Maintenance.

  • Set low temperatures to avoid burns and injuries.
  • Set reduced flows to avoid safety hazards and prevent instrument damage.
  • Turn off the GC and disconnect from power.
  • Vent a mass selective detector (MSD) if present.
  • Maintenance Mode ends when the GC is turned off.
  • If not using built-in features, first cool all heated zones, then turn off GC and unplug power cord.
pp. 14
Automated Maintenance Procedures

The GC provides automated procedures that walk through routine maintenance tasks, including cool downs, leak tests, EMF resets, and blank runs. These are accessible from the touchscreen and browser interface under Maintenance > [device] > Perform Maintenance. Procedures require an idle GC and will not run if the GC is not idle.

  • Automated procedures include cool downs, leak tests, EMF resets, and blank runs.
  • Access via Maintenance, select device, then Perform Maintenance.
  • Procedures require an idle GC.
  • Data systems can prevent automated confirmation runs (e.g., blank diagnostic runs) after replacing certain parts; this is normal.
  • Leak checks and pressure checks will run during maintenance since they do not start a run.
pp. 15
Viewing Status Data

The GC can display signal and other data useful during maintenance, including instrument status, instrument status details, and module status details.

  • Instrument status: readiness state, running state.
  • Instrument status details: not ready modules, error modules, warnings.
  • Module status details: actual/setpoint pairs for each module, quick on/off for certain modules, firmware version and instrument IP address.
pp. 16
Finding a Replacement Part Number

Agilent provides the Parts Finder tool to locate replacement and consumable part numbers. It is installed with Agilent data systems or available on the Agilent GC and GC/MS User Manuals & Tools DVD. Parts can be navigated graphically by location in the GC.

  • Parts Finder tool helps locate replacement and consumable part numbers.
  • Installed with Agilent data systems or available on the User Manuals & Tools DVD.
  • Navigate to parts graphically based on location in the GC.
  • Part numbers are also included in this manual.
pp. 17-20
Removing Covers

This section describes how to remove the detector top cover, pneumatics cover, and electronics cover as needed for routine maintenance. Only these covers should be removed; removing other covers can compromise safety features.

  • Detector top cover protects detectors, valve box, and valve assembly.
  • Electronics cover may need to be raised for NPD maintenance; steps differ depending on whether an FPD+ is installed.
  • Always turn off GC and unplug power cord before removing covers.
  • Do not force covers; could break plastic parts.
  • Only covers listed in this chapter should be removed.
pp. 24-25
Consumables and Parts for General GC Maintenance

Table 2 lists consumable parts for the hydrogen sensor module accessory and general GC maintenance, including calibration gas cylinders, regulators, fittings, and Gas Clean filters.

  • Calibration gas cylinder, 2 percent hydrogen gas in 17 L compressed air: 5190-6890
  • Cylinder stand with locking screw: 1400-3583
  • Pressure regulator, with gage: G3440-80153
  • Fittings kit, 1/8-inch brass, 20/pk: 5080-8750
  • Regulator, 2-stage, brass body, stainless steel diaphragms, 125 psi max, CGA350, hydrogen, argon/methane, with 1/8-inch fitting: 5183-4642
  • Regulator, 2-stage, brass body, stainless steel diaphragms, 125 psi max, CGA346, air, with 1/8-inch fitting: 5183-4641
  • Regulator, 2-stage, brass body, stainless steel diaphragms, 125 psi max, CGA580, helium, argon, nitrogen, 1/8-inch fitting: 5183-4644
  • Gas Clean connecting unit, 1-position, 1/4-inch: CP7980
  • Gas Clean filter oxygen, 1/pk: CP17970
  • Gas Clean filter moisture, 1/pk: CP17971
  • Gas Clean filter charcoal, 1/pk: CP17972
pp. 27-28
To Install the Hydrogen Sensor Calibration Gas Cylinder

Procedure for replacing an existing hydrogen calibration gas cylinder. Only Agilent calibration gas (2% hydrogen in air) should be connected. Calibration is initiated via the browser interface and requires adjusting the pressure regulator to achieve approximately 30 mL/min flow.

  • Connect only Agilent calibration gas (2% hydrogen in air) to the hydrogen sensor fitting.
  • Cool oven and detector to < 40 °C before calibration.
  • Initiate calibration: Settings > Calibration > H2 Sensor > Start Calibration.
  • Adjust pressure regulator until flow rate is approximately 30 mL/min (+/- 3 mL).
  • Calibration cycle takes approximately 5 minutes total.
  • If calibration fails, check calibration gas type/level, leaks, restrictions, and flow rate.
pp. 30-31
Consumables and Parts for Columns

Tables 3 and 4 list nuts, ferrules, hardware, and hangers for capillary columns, organized by column inner diameter.

  • Ferrule, Vespel/graphite, 0.8-mm id for 0.45-mm and 0.53-mm capillary columns: 5062-3512 (10/pk)
  • Ferrule, graphite, 1.0-mm id for 0.53-mm capillary columns: 5080-8773 (10/pk)
  • Ferrule, Vespel/graphite, 0.5-mm id for 0.32-mm capillary columns: 5062-3514 (10/pk)
  • Ferrule, graphite, 0.5-mm id for 0.1-mm, 0.2-mm, 0.25-mm, and 0.32-mm capillary columns: 5080-8853 (10/pk)
  • Column nut, finger-tight (for 0.53-mm columns): 5020-8293
  • Column nut, finger-tight (for .100- to .320-mm columns): 5020-8292
  • Ferrule, no-hole (testing): 5181-3308 (10/pk)
  • Column cutter, ceramic wafer: 5181-8836 (4/pk)
  • Column hanger: 1460-1914
  • Capillary column clip kit, for 7-inch column basket: G1530-61580
pp. 32-39
Capillary Column Installation and Conditioning

Procedures for installing a capillary column hanger and clips, conditioning a capillary column, cutting a loop, and reversing a column to bake out contaminants. Conditioning requires setting a minimum velocity of 30 cm/s, oven at 120 °C, then programming to the column maximum temperature at 10 to 15 °C/min and holding for 30 minutes.

  • Do not use hydrogen as carrier for conditioning; explosion hazard.
  • Set inlet and oven temperature to < 40 °C before conditioning.
  • Set minimum velocity of 30 cm/s or as recommended by column manufacturer.
  • Set oven to 120 °C and let gas flow for 15 to 30 minutes to remove air.
  • Program oven from 120 °C to column maximum temperature at 10 to 15 °C/min; hold at maximum for 30 minutes.
  • For bakeout after reversing column: set inlet to 300 °C or 25 °C above normal operating temperature; set oven 25 °C above method final oven temperature; bake for 30 minutes.
  • For Split/Splitless inlets, select split mode and set split vent flow to 200 mL/min during bakeout.
  • Purge column with carrier flow for at least 10 minutes before heating oven.
pp. 42-44
Consumables and Parts for the Split/Splitless Inlet

Tables 5, 6, and 7 list inlet liners, nuts, ferrules, septa, O-rings, and other parts for the Split/Splitless Inlet.

  • Split liner, low-pressure drop, glass wool, single taper, 870 µL, deactivated: 5183-4647
  • Splitless liner, single taper, glass wool, 900 µL, deactivated: 5062-3587
  • Splitless liner, single taper, no glass wool, 900 µL, deactivated: 5181-3316
  • Ultra Inert, low pressure drop, glass wool liner (split or splitless): 5190-2295
  • 11-mm septum, high-temperature, low-bleed, 50/pk: 5183-4757
  • Nonstick fluorocarbon liner O-ring (up to 350 °C), 10/pk: 5188-5365
  • Graphite O-ring for split liner (above 350 °C), 10/pk: 5180-4168
  • Graphite O-ring for splitless liner (above 350 °C), 10/pk: 5180-4173
  • Gold-plated seal (standard application): 5188-5367
  • Gold-plated seal with cross (high split flows): 5182-9652
pp. 46-49
To Install a Capillary Column with the Split/Splitless Inlet

Procedure for installing a capillary column using the GC maintenance wizard. Includes verifying correct glass liner, placing column on hanger, installing septum/nut/ferrule, tightening the column nut an additional 1/4 to 1/2 turn, configuring and conditioning the column, and installing into the detector.

  • Launch wizard: Maintenance > Column > Perform Maintenance > Install Column > Start Maintenance.
  • Verify correct glass liner is installed.
  • Tighten column nut an additional 1/4 to 1/2 turn with a wrench so column cannot be pulled from fitting with gentle pressure.
  • Wizard performs leak & restriction tests and automatically resets maintenance counters.
  • After installation, establish carrier gas flow and purge as recommended by column manufacturer.
  • For FID or FPD+, immediately turn off flame; for NPD, immediately turn off bead.
  • After GC becomes ready, wait 10 minutes then ignite detector flame or bead.
  • Allow oven, inlet, and detector to equilibrate at operating temperature, then retighten fittings.
pp. 195
To Replace the NPD Bead Assembly

Procedure for replacing the NPD bead assembly, including closing covers, exiting maintenance mode, restoring gas flows, heating the detector in stages, checking leakage current, and resetting the bead counter.

  • All covers must be closed to get a stable NPD baseline.
  • Exit maintenance mode: Maintenance > Instrument > Perform Maintenance > Maintenance Mode > Finished.
  • Heat detector to 150 °C and hold for about 15 minutes, then increase to 250 °C and hold for 15 minutes.
  • Increase temperature to operating value (310 to 320 °C recommended); allow 15 minutes for equilibration.
  • Check NPD leakage current; if > 2.0 pA, verify bead installation or see Troubleshooting manual.
  • Start Adjust offset process; an offset of 25 to 30 pA is sufficient for most applications.
  • Higher offset may shorten bead life.
  • Reset the bead counter.

Key Integration, Operating & Safety Requirements

Specification / Condition Engineering Implication Applies To
Hydrogen sensor calibration gas flow must be adjusted to approximately 30 mL/min (+/- 3 mL) using the pressure regulator on the calibration gas cylinder. p.28 ↗ Ensures proper calibration of the hydrogen sensor; incorrect flow can cause calibration failure. 8860 GC with hydrogen sensor module
Only Agilent calibration gas (2% hydrogen in air) should be connected to the hydrogen sensor fitting; higher hydrogen concentrations can create a dangerous condition or damage the sensor. p.27 ↗ Prevents explosion hazard and sensor damage. 8860 GC with hydrogen sensor module
Cool oven and detector to < 40 °C before hydrogen sensor calibration. p.27 ↗ Prevents burns during calibration procedure. 8860 GC with hydrogen sensor module
For capillary column conditioning, set a minimum carrier gas velocity of 30 cm/s, oven at 120 °C, then program to column maximum temperature at 10 to 15 °C/min and hold for 30 minutes. p.35 ↗ Proper conditioning removes contaminants and ensures column performance. Capillary columns
Do not use hydrogen as carrier gas for column conditioning; it could vent into the oven and present an explosion hazard. p.34 ↗ Safety requirement during conditioning. Capillary columns
When installing a capillary column in the Split/Splitless inlet, tighten the column nut an additional 1/4 to 1/2 turn with a wrench so the column cannot be pulled from the fitting with gentle pressure. p.48 ↗ Ensures a leak-free connection without over-tightening. Split/Splitless Inlet
After installing a capillary column, for FID or FPD+ immediately turn off the flame; for NPD immediately turn off the bead. After GC becomes ready, wait 10 minutes then ignite detector flame or bead. p.48 ↗ Prevents damage to detector and ensures stable ignition. FID, FPD+, NPD
For NPD bead replacement, heat detector to 150 °C for about 15 minutes, then to 250 °C for 15 minutes, then to operating value (310 to 320 °C recommended) and allow 15 minutes for equilibration. p.195 ↗ Proper conditioning of new bead ensures stable baseline and sensitivity. NPD
NPD leakage current should be ≤ 2.0 pA; if > 2.0 pA, verify bead installation or consult Troubleshooting manual. p.195 ↗ Indicates proper bead installation and detector condition. NPD
For NPD Adjust offset, an offset of 25 to 30 pA is sufficient for most applications; higher offset may shorten bead life. p.195 ↗ Balances sensitivity and bead lifetime. NPD
When reversing a column for bakeout, set inlet temperature to 300 °C or 25 °C above normal operating temperature, and set oven 25 °C above method final oven temperature; bake for 30 minutes. p.39 ↗ Removes contaminants from inlet and column without exceeding column temperature limits. Capillary columns
For Split/Splitless inlets during bakeout, select split mode and set split vent flow to 200 mL/min. p.39 ↗ Ensures contaminants are flushed through split vent. Split/Splitless Inlet
Purge column with carrier flow for at least 10 minutes before heating oven during bakeout. p.39 ↗ Removes air and prevents column damage. Capillary columns
All covers must be closed to get a stable NPD baseline. p.195 ↗ Ensures proper thermal and electrical shielding for NPD. NPD
Automated maintenance procedures require an idle GC and will not run if the GC is not idle. p.14 ↗ Ensure GC is idle before starting automated maintenance. 8860 GC
Data systems can prevent automated confirmation runs (e.g., blank diagnostic runs) after replacing certain parts; this is normal behavior. p.14 ↗ Manual confirmation run may be needed if data system disables local run features. 8860 GC with data system
Maintenance Mode ends when the GC is turned off. p.13 ↗ If GC is turned off during maintenance, Maintenance Mode must be re-entered. 8860 GC
Only covers listed in the Removing Covers chapter should be removed; removing other covers can compromise safety features. p.17 ↗ Prevents personal injury or instrument damage. 8860 GC
Use electronic flow meter(s) or bubble meter(s) capable of calibrated measurements at 1, 10, and 100 mL/min flow ranges for maintenance. p.11 ↗ Ensures accurate flow measurements during maintenance. 8860 GC
Nonstick fluorocarbon liner O-ring is for temperatures up to 350 °C; graphite O-rings are for temperatures above 350 °C. p.43 ↗ Select correct O-ring material based on operating temperature. Split/Splitless Inlet

Faults & Troubleshooting

Code / Symptom Possible Cause Remedy Note
Hydrogen sensor calibration failed p.28 ↗ Incorrect calibration gas type, empty or low tank, leaks in calibration gas supply, restrictions in supply tubing, or calibration gas not flowing at expected rate of 30 mL/min. Check calibration gas type and level, check for leaks, check for restrictions, adjust pressure to achieve 30 mL/min flow. Settings > Service Mode > H2 Sensor > Calibration
NPD leakage current > 2.0 pA p.195 ↗ Improper bead installation or other issue. Verify bead installation or see the Troubleshooting manual. After NPD bead replacement