1 Refer to 1260 Infinity II Isocratic Pump User Manual for complete pump specifications 2 The technical specifications have been determined by experimentation and reflect the total system performance. The values quoted have been calculated from analyses performed in THF at 40 °C and in TCB at 160 °C.
Table 1 Physical Specifications 1260 Infinity II High Temperature GPC System
Type
Specification
Comments
Line voltage
230 VAC (±10 %)
Single phase with earth
Power consumption
3300 VA, 2310 W/7900 BTU
Line frequency
50 or 60 Hz (±5 %)
Size (width × depth × height)
900 mmx 530 mm x 620 mm
Packaged Size (width × depth × height)
930 mm x 760 mmx 860 mm
Weight
136 kg (300 lbs.)
Ambient operating temperature
15 – 30 °C (59 – 86 ° F)
At constant temperature
Ambient non-operating temperature
-40 – 70 °C (-4 – 158 ° F)
Operating altitude
Up to 2000 m (6562 ft.)
Non-operating altitude
Up to 4600 m (15091ft.)
Humidity
<80 % at 25 – 30 °C
Safety standards: IEC, CSA, UL
Installation category II, Pollution degree 2
For indoor use only.
Figure 5 Standard Oven Configuration
Number
Component
1
Oven Fan with Heater
2
Temperature Sensors
3
Differential Refractive Index Detector
4
Sample Needle Assembly
5
Column Mounts
6
Injection Valve
7
Vapour Sensor
Agilent 1260 DataStream
No.
Description
No.
Description
1
Channel start button
2
Status LEDs
3
Channel analogue input
4
CTR connector
6
PL-CAN connector (not used)
7
Aux I/O connector
8
Power Switch
9
Voltage selector
5
RS232 connector
10
Power inlet
Figure 4 Instrument rear view and connections
Number
Connection
1
Refractive Index signal output ±1V FSD
2
Auxiliary I/O - Pump stop
3
Oven Power Cable connector
4
Injection synchronisation to Agilent DataStream
5
Serial connection to PC (Instrument Control software)
6
LAN connection to Agilent pump
7
Oven RI link cable connector
8
Oven Digital link cable connector
Table 7 Auto-ignition temperatures of common GPC solvents with ELS detection
Solvent
Auto-ignition Temperature (°C)
DMSO
215
Dibutoxymethane
220
Tetrahydrofuran
321
Dimethylformamide
445
Toluene
480
Acetone
465
1,2,4 Trichlorobenzene (TCB)
570
m-Cresol
558
Chloronaphthalene
558
Table 3 Error overview
Type
GUI Error Message
Conditions
Response Time (Timeout, Retries)
LED
Beeper
Resulting Mode
Notes
Oven
TEMPERATURE "Oven temperature has become unstable"
Temperature setpoint reached, however subsequently increased to > 5 °C above setpoint
- (-,-)
Flash Red
NO
-
Not STBY
FAN
“The fan is no longer operating correctly"
Frequency < 10 kHz
5 s (5 s)
Flash Red
YES
STBY
DOOR
"Illegal opening of the oven door has occurred"
I/O state change
- (-,-)
Flash Orange
YES
-
Heaters OFF while door open
V/F
"Temperature sensor is not responding correctly"
Temp = 0 °C
- (-,-)
Flash Red
YES
STBY
SENSOR
"The vapour sensor is not responding correctly"
Open/Short circuit
5 s (5 s)
Flash Red
YES
STBY
OVEN VAPOUR
"Spillage may have occurred within the oven"
Voltage < set threshold
- (-, -)
Flash Orange
YES
-
2 min (-, -)
Flash Red
YES
STBY
Table 3 Error overview
Type
GUI Error Message
Conditions
Response Time (Timeout, Retries)
LED
Beeper
Resulting Mode
Notes
RI
AZFAIL
"Attempts to autozero the RI have failed"
Timeout or False Zero
20 min (10 min, 2)
Flash Orange
NO
-
Not while initializing. Injection sequence is aborted.
PURGE
"The purge valve is not functioning correctly"
Timeout
5 s (5 s)
Flash Orange
NO
-
Not while initializing. Injection sequence is aborted.
Autosampler
INITIALIZATION
"Failed to initialize autosampler"
Command failed (response from autosampler)
- (-, 0)
Flash Red
NO
-
MOVE
"Failed to move carousel correctly"
- (-, 3)
Flash Red
NO
-
VIALLOWER
"Failed to lower vial from the oven"
- (-, 0)
Flash Red
NO
-
VIALLIFT
"Failed to lift vial from the carousel"
- (-, 0)
Flash Red
NO
-
SYRINGE
"Failed to move syringe"
- (-, 0)
Flash Red
NO
-
LOAD
"Failed to switch valve to load position"
- (-, 3)
Flash Red
NO
-
INJ
"Failed to switch valve to inject position"
- (-, 3)
Flash Red
NO
-
Table 3 Error overview
Type
GUI Error Message
Conditions
Response Time (Timeout, Retries)
LED
Beeper
Resulting Mode
Notes
CAROUSEL LOCK
"Failed to lock carousel after being closed"
Command failed during autosampler run
- (-, 0)
Flash Red
NO
-
AUTOCOMMS
"The autosampler is no longer responding"
Packet errors/ no ACK response from autosampler
1.5 s (500 ms, 3)
Flash Red
NO
no command success response
variable,
(variable, 0)
Flash Red
NO
Pump
PUMPCOMMSERR
“The pump is no longer responding.”
No response, or invalid response from pump.
~30s (~10s, 3)
Flash Red
YES
STANDBY
PUMPBADTYPEID
"Pump has wrong type id"
Pump is not of the type G7110B.
- (-, -)
Flash Red
YES
STANDBY
PUMPERRCODE
An error code, along with an appropriate message. The message will vary based on the error code.
Pump is in an error state.
- (-, -)
Flash Red
YES
STANDBY
See Pump General Error Messages and Pump Module Error Messages
SEQABORTPUMPNRDY
“Pump became not ready. See error log.”
Pump became not-ready while the autosampler was running an injection sequence.
- (-, -)
Flash Orange
YES
Injection sequence is aborted.
MCGV Fuse
Probable cause
Suggested actions
1 Defective gradient valve.
Restart the quaternary pump. If the error message appears again, exchange the gradient valve.
2 Defective connection cable (front panel to main board).
Please contact your Agilent service representative.
3 Defective main board.
Please contact your Agilent service representative.
Error ID: 2043; Valve Fuse 0: Channels A and B; Valve Fuse 1: Channels C and D. The gradient valve in the quaternary pump has drawn excessive current causing the electronic fuse to open.
Engineering Facts
14 facts · every claim linked to a page
Installation
The instrument requires a dedicated 230 VAC single-phase earthed supply with an RCCB, so site power must be planned before delivery.
Electrical infrastructure and breaker selection must be confirmed during site preparation.
Source
“The mains supply should be fitted with an RCCB (Residual Current operated Circuit-Breaker).”
Installation
The power plug must remain reachable at all times so the unit can be disconnected in an emergency.
Bench layout must leave free access behind the instrument for safe disconnection.
Source
“Make sure the power connector of the instrument can be easily reached and unplugged.”
Installation
Only the Agilent power cord matched to the region may be used, and the outlet must be grounded.
Prevents electric shock, short circuit and EMC non-compliance at commissioning.
Source
“Never use a power cord other than the Agilent Technologies power cord designed for your region.”
Installation
The bench must accommodate the footprint and the upward-opening oven, which needs 97 cm clear height.
Overhead obstructions above the bench must be checked before installation.
Source
“The oven opens upwards to a full height of 97 cm (38 in).”
Selection
The system draws 3300 VA / 2310 W, which drives the electrical supply and heat-load sizing.
Lab power capacity and HVAC load must be verified against this consumption.
Source
“3300 VA, 2310 W/7900 BTU”
Selection
The 136 kg weight and 900 x 530 x 620 mm size dictate floor loading and bench strength requirements.
Bench or floor must be rated for the mass and dimensions of the main module.
Source
“136 kg (300 lbs.)”
Operation
The oven reaches 220 °C with better than 0.05 °C/h stability, enabling high-temperature GPC at 160 °C.
Confirms the thermal envelope needed for TCB-based polyolefin analysis.
Source
“Ambient to 220
° C Temperature Stability
<0.05 ° C/h”
Operation
The autosampler carousel holds 40 x 2 mL vials, with position zero reserved as the index vial.
Sample throughput planning must account for only 39 usable injection positions.
Source
“The carousel will accommodate 40 x 2 mL sample vials, 39 of these (positions
1 to 39) are used for sample injection.”
Operation
An empty, clean, capped vial must always sit in position zero or the autosampler cannot initialize and waste disposal fails.
Missing index vial blocks initialization and risks polymer precipitation in the loop.
Source
“It is therefore important to keep an empty vial fitted
with a clean, shiny cap in the zero position of the carousel at all times.”
Operation
Solvents must be degassed and filtered to 0.45 µm and be compatible with the column before use.
Improper solvent preparation causes baseline noise, blockages and column damage.
Source
“should be thoroughly degassed, filtered (0.45 µm) and fully compatible with the
column(s).”
Safety
Solvents with auto-ignition temperatures below 190 °C must not be used in the system.
Solvent selection at method development must respect this fire-safety limit.
Source
“Do not use solvents with auto-ignition temperatures <190°C with the HT GPC System”
Operation
Ambient lab temperature must exceed the solvent melting point, e.g. above 19 °C for TCB, before pumping.
Cold labs can freeze TCB and block the pump; HVAC setpoint must be verified.
Source
“the ambient temperature must exceed 19°C for 1,2,4-trichlorobenzene before
the solvent can be pumped”
Maintenance
Column changes require cooling to 40 °C, setting idle/low flow to 0.2 mL/min, then STANDBY before opening the oven.
Defines the safe, leak-checked procedure for column exchange at high temperature.
Source
“Change the Idle and Low flow rates to 0.2 mL/min and the Idle
temperature to 40 °C.”
Maintenance
After column change the DRI reference cell must be purged for at least 20 minutes before use.
Insufficient purging leaves air bubbles and unstable RI baseline.
Source
“Purge the DRI reference cell for at least 20 min using the Control
software.”