Contenido

  1. Overview pp. 10-15
    Product description and operating principle of the autosampler
  2. Overview pp. 19-21
    Description of the hydraulic box assembly
  3. Procedure pp. 22-24
    Instructions for leak and waste handling
  4. Spec pp. 25-27
    Physical site and environmental requirements
  5. Spec pp. 28-30
    Dimensions, weight and physical specifications
  6. Spec pp. 34-36
    Specifications of the sample thermostat option
  7. Procedure pp. 37-75
    Operating procedures for the autosampler and thermostat
  8. Spec pp. 79-90
    Tubing and fitting compatibility with solvents
  9. Procedure pp. 91-99
    Setup and preparation of the sampler
  10. Procedure pp. 100-111
    Injector program and carry-over optimization
  11. Procedure pp. 112-120
    Maintenance commands and part replacement
  12. pp. 121-123
    List of step commands for the module
  13. Procedure pp. 124-156
    Error messages and troubleshooting procedures
  14. Overview pp. 157-162
    Overview of maintenance procedures and intervals
  15. Procedure pp. 163-207
    Replacing needle, rotor seal, head, gripper and pump
  16. pp. 208-219
    Main assemblies and accessory kits overview
  17. pp. 220-231
    Cable types, connectors and identification
  18. pp. 232-249
    Firmware description, electrical and special interfaces
  19. Procedure pp. 250-273
    TCP/IP and LAN parameter configuration
  20. Warning pp. 274-288
    Safety warnings and sound emission information
  21. p. 256 table, 3 rows
  22. p. 262 table, 6 rows
  23. p. 29 table, 2 rows
  24. p. 35 table, 4 rows
  25. p. 30 table, 9 rows
  26. p. 31 table, 8 rows
  27. p. 32 table, 12 rows
  28. p. 33 table, 4 rows
  29. p. 34 table, 11 rows
  30. p. 255 table, 3 rows

Table 26 Using Default Parameters

ParameterDefault Parameter
IP address:192.168.254.11
Subnet Mask:255.255.255.0
Default Gatewaynot specified

Table 29 Telnet Commands

ValueDescription
?displays syntax and descriptions of commands
/displays current LAN settings
ip <x.x.x.x>sets new ip address
sm <x.x.x.x>sets new subnet mask
gw <x.x.x.x>sets new default gateway
exitexits shell and saves all changes

Table 2 Performance Specifications G7157A

TypeSpecificationCommentsMethod/Conditions
Settable injection range0.1 – 900 µL, up to 3600 µL with seat extension
Injection precision1 µL: <5 % 5 µL: <2 % 10 µL, 50 µL: <1 % 500 – 3600 µL: <0.25 %Measured caffeineSample: Caffeine in Water/ACN (98/2) for 1 – 50 µL Injection Volume: • Column: Agilent Zorbax SB-C18, 4.6 x 50 mm, 5 µm (PN: 846975-902) • Eluent: Water/ACN (90/10) isocratic, premixed pumping Channel A/B (50/50) • Flow rate: 5.0 mL/min • Temperature: RT • Detection: 273/4 nm Ref. 360/100 nm, 10 Hz • Needle wash: 5 s in flush port with Water/ACN (90/10) • Draw speed: 100 µL/min • Eject speed: 3000 µL/min • Wait time after draw: 5 s for 500 – 3600 µL Injection Volume: • Column: Agilent 5 Prep-C18, 21.2 x 50 mm, 5 µm (PN: 446905-102) • Eluent: Water/ACN (85/15) isocratic • Flow rate: 30.0 mL/min • Temperature: RT • Detection: 273/4 nm Ref. 360/100 nm, 10 Hz • Needle wash: 5 s in flush port with Water/ACN (85/15) • Draw speed: 900 µL/min • Eject speed: 3000 µL/min • Wait time after draw: 5 s • for 2000 – 3600 µL: seat extension (PN: G7157-68711)

Table 6 Performance Specifications for the Sample Thermostat (G7167-60101 and G7167-60201)

TypeSpecifications
Operating principleHigh performance, low-energy consumption micro-compressor based cooler with natural R600a coolant (Isobutane max. 0.030 kg), user-upgradable
Temperature rangefrom 4 °C to 40 °C
Temperature settablefrom 4 °C to 40 °C in 1 ° increments
Temperature accuracy (<25 °C, <50 % r.H.)2 °C to 6 °C at a setpoint of 4 °C

Table 2 Performance Specifications G7157A

TypeSpecificationComments
Pressure range0 – 40 MPa (0 – 400 bar, 0 – 5801.51 psi)
Sample viscosity range0.2 – 5 cP
Sample capacity132 x 2 mL vial (two trays default)
100 x 2 mL vial (two classic trays optional)
36 x 6 mL vials (two trays optional)
Carryover<0.005 % (50 ppm)
Method• Column: Agilent 5 Prep-C18, 21.2 mm x 50 mm, 5 µm (446905-102) • Eluent: Water/ACN (85/15) isocratic • Flow rate: 30.0 mL/min • Temperature: RT • Detection: 273/4 nm Ref. 360/100 nm, 10 Hz • Injection volume: 500 µL • Needle wash: 10 s in flush port with Water/ACN (85/15) • Draw speed: 900 µL/min • Eject speed: 3000 µL/min • Wait time after draw: 5 s
Sample• Caffeine 5 g/L in Water /ACN (98/2)
Method/Conditions

Table 2 Performance Specifications G7157A

TypeSpecificationComments
Injection cycle time<60 s for injection volume 900 µL
Method• Column: Restriction capillary (0.25 x 4000 mm; PN: 5065-4493) • Eluent: Water/ACN (85/15) isocratic • Flow rate: 5.0 mL/min • Temperature: RT • Injection volume: 900 µL • Needle wash: no needle wash • Draw speed: 9000 µL/min • Eject speed: 9000 µL/min • Wait time after draw: 1.2 s • Overlapped injection mode: prefetch vialMethod/Conditions
SampleWater
Instrument ControlLC & CE Drivers A.02.17 or above Instrument Control Framework (ICF) A.02.04 or above Instant Pilot (G4208A) with firmware B.02.22 or above Lab Advisor B.02.10 or aboveFor details about supported software versions refer to the compatibility matrix of your version of the LC and CE Drivers
CommunicationsController-area network (CAN), Local Area Network (LAN) ERI: ready, start, stop and shut-down signals
Maintenance and safety-related featuresExtensive diagnostics, error detection and display with Agilent Lab Advisor software Leak detection, safe leak handling, leak output signal for shutdown of pumping system, and low voltages in major maintenance areas
GLP featuresEarly maintenance feedback (EMF) for continuous tracking of instrument usage with user-settable limits and feedback messages. Electronic records of maintenance and errors.
HousingAll materials recyclable.

Table 3 Physical Specification of the Sample Cooler

TypeSpecificationComment
Weight< 6 kg (< 13.2 lbs)
Dimensions (height × width × depth)205 x 340 x 370 mm (8.1 x 13.4 x 14.6 inches)
Refrigerant gasHFC-134a (0.042 kg)Ozone depletion potential (ODP) = 0
Supply voltage24 VDC
Current10 A max.
Ambient operating temperature4 – 40 ° C (39 – 104 ° F)
Ambient non-operating temperature-40 – 70 ° C (-40 – 158 ° F)
Humidity< 95 % r.h. at 40 °C (104 °F)Non-condensing
Operating altitudeUp to 3000 m (9842 ft)
Safety standards:IEC, EN, CSA, ULInstallation category II, Pollution degree 2
For indoor use only
ISM ClassificationISM Group 1 Class BAccording to CISPR 11

Table 4 Performance Specifications of the Sample Cooler

TypeSpecifications
Operating principleHigh performance, low-energy consumption micro-compressor based cooler with ozone-friendly HFC-134a coolant (42 g), user-upgradable.
Temperature rangefrom 4 °C to 5 °C below ambient
Temperature settablefrom 4 – 40 °C in 1 ° increments
Temperature accuracy (<25 °C, <50 % r.H.)2 °C to 6 °C at a setpoint of 4 °C

The Agilent Infinity II Sample Cooler is not available for trade sales anymore and has been replaced by the Agilent InfinityLab Sample Thermostat.

Table 5 Physical Specifications of the Sample Thermostat (G7167-60101 and G7167-60201)

TypeSpecificationComments
Weight<6 kg (< 13.2 lbs)
Dimensions (height x width x depth)205 x 340 x 370 mm (8.1 x 13.4 x 14.6 inches)
Refrigerant gasR600a (max. 0.030 kg)Ozone depletion potential (ODP) =0 Global warming potential (GWP) =3
Supply voltage24VDC
Current10 A max.
Ambient operating temperature4 °C to 40 °C (39 °F to 104 °F)For sample cooling, ambient temperature ≥10 °C
Ambient non-operating temperature-40 °C to +70 °C (-40 °F to +158 °F)
Humidity< 95 % r.h. at 40 °C (104 °F)Non-condensing
Operating altitudeUp to 3000 m (9842 ft)
Safety standards: IEC, EN, CSA, ULPollution degree 2For indoor use only
ISM ClassificationISM Group 1 Class BAccording to CISPR 11

Table 25 Initialization Mode Switches

SW1SW2SW3SW4SW5SW6Init Mode
000000Use Default IP Address
000010Use Stored IP Address
000100Use DHCP

Note: The setting ‘0’ (down) is essential.

Engineering Facts

12 facts · every claim linked to a page
Installation

The module requires additional clearance on the sides and rear for air circulation and electrical connections.

Ensures adequate ventilation and access for cables during bench installation.

Source
“It needs an additional 2.5 cm (1.0 inches) of space on either side and approximately 8 cm (3.1 inches) in the rear for air circulation and electric connections.”
Installation

The module must be operated horizontally, especially with a Sample Cooler or Thermostat, and leveled using a bubble level.

Prevents mechanical issues and ensures proper operation of the sampler and cooler/thermostat.

Source
“The module should be operated in a horizontal position, especially if a Sample Cooler or Sample Thermostat is installed. Use a bubble level to check the leveling of the sampler.”
Safety

The module is partially energized even when switched off if the power cord is plugged in.

Highlights electrical shock hazard during maintenance or service.

Source
“The module is partially energized when switched off, as long as the power cord is plugged in.”
Installation

The power connector must be easily reachable and unpluggable in case of emergency.

Ensures compliance with safety requirements for emergency disconnection.

Source
“Make sure the power connector of the instrument can be easily reached and unplugged.”
Safety

Only the power cord shipped by Agilent should be used with the instrument.

Prevents personal injury or equipment damage from incompatible power cords.

Source
“Never use a power cord other than the one that Agilent shipped with this instrument.”
Operation

For injection volumes below 2 µL, degassed mobile phases are required to avoid gas bubbles affecting precision.

Critical for achieving accurate low-volume injections.

Source
“Use degassed mobile phases for best injection-volume precision with injection volumes < 2 µL.”
Operation

The default draw speed is 100 µL/min, but lower speeds are recommended for viscous samples to avoid air bubbles.

Improves injection volume precision when handling viscous samples.

Source
“The default draw speed is 100 µL/min. This speed is suitable for the majority of applications, however, when using viscous samples, set the draw speed to lower speed for optimum results.”
Operation

The default eject speed is 400 µL/min; higher speeds shorten the injection cycle but should be avoided for viscous samples.

Optimizes cycle time while avoiding issues with viscous samples.

Source
“The default eject speed setting is 400 µL/min. When using large injection volumes, setting the eject speed to a higher value speeds up the injection cycle by shortening the time the metering unit requires to eject solvent at the beginning of the injection cycle (when the piston returns to the home p”
Maintenance

The peristaltic pump cartridge is a replaceable unit; the tubing inside is not replaceable.

Informs maintenance planning and spare parts ordering.

Source
“The peristaltic pump cartridge is a replaceable unit. The tubing inside the pump is not replaceable.”
Maintenance

When replacing the rotor seal, do not over-tighten the stator head screws; sealing force is automatic.

Prevents damage to the valve during maintenance.

Source
“Do not over-tighten the screws. The screws hold the assembly together and do not affect the sealing force. The sealing force is automatically set as the screws close the stator head against the valve body.”
Installation

Transporting the sampler with a Sample Cooler/Thermostat installed is only allowed for short distances; for longer distances, separate the units.

Avoids mechanical damage during relocation.

Source
“Transporting the sampler with a Sample Cooler/Sample Thermostat installed is only allowed for short distances. For longer distances, you must separate the units and send them independently.”
Safety

The leak sensor tubing is made of PVDF, which is incompatible with DMF.

Prevents solvent compatibility issues that could damage the leak sensor.

Source
“The tubing of the leak sensor is made of PVDF (polyvinylidene fluoride), which is incompatible with the solvent DMF (dimethylformamide).”