1260 Infinity III Prime Online LC and 1290 Infinity III Bio Online LC System Manual

Contenido

  1. Overview pp. 5-10
    Product descriptions and features of Online LC systems
  2. Overview pp. 11-28
    Descriptions of individual modules and detectors
  3. Procedure pp. 29-39
    Installing modules, network integration, stack configuration, tubing
  4. Procedure pp. 40-45
    Waste concept, waste guidance, leak sensor handling
  5. Config. pp. 46-73
    LAN, instrument, software configuration and notifications setup
  6. Procedure pp. 74-107
    Preparing pump, solvents, runs and experiments
  7. Procedure pp. 108-124
    Checkout method setup and procedure in monitoring software
  8. Ordering pp. 125-145
    Kits, capillaries, fittings, bottles and additional parts
  9. p. 109 table, 8 rows
  10. p. 110 table, 3 rows
  11. p. 111 table, 4 rows
  12. p. 38 table, 12 rows
  13. p. 39 table, 5 rows
  14. p. 128 table, 14 rows
  15. p. 133 table, 5 rows
  16. p. 139 table, 16 rows

Table 4: Checkout method parameter settings Flexible Pump (G7104C), Bio High-Speed Pump (G7132A) and Bio Flexible Pump (G7131A)

ParameterValue
Flow0.8 mL/min
Solvent AWater
Solvent BACN
CompressibilityUse solvent types
Composition5 %B (ACN)
Composition95 %A (Water)
Stoptime5 min
Pressure Limit800 bar

Table 5: Checkout method parameter settings Online Sample Manager (G3167A/B)

ParameterValue
Minimum StrokeAutomatic
Timetable2.5 min 95 %B
Posttime1 min

Table 8: Checkout method parameter settings Variable Wavelength Detector (G7114A/B)

ParameterValue
Wavelength254 nm
Peakwidth40 Hz
Stoptimeas pump
AutobalancePrerun

Table 2: Capillary connections for 1260 Infinity III systems

p/nFromTo
G7120-60007 (Bottle Head Assembly, long (1730 mm))Solvent BottleInfinity III Pump
5500-1246 (Capillary ST 0.17 mm x 500 mm SI/SI)PumpSampler
5500-1217 (Capillary, ST, 0.17 mm x 900 mm SI/SX)PumpVialsampler with ICC
5500-1246 (Capillary ST 0.17 mm x 500 mm SI/SI)MultisamplerMCT Valve/Heat Exchanger
5500-1252 (Capillary, ST, 0.17 mm x 400 mm SL/SL)VialsamplerMCT Valve/Heat Exchanger
5500-1240 (Capillary ST 0.17 mm x 105 mm SL/SL)VialsamplerICC Heat Exchanger
5500-1250 (Capillary, ST, 0.17 mm x 120 mm SL/SL, long socket)ICC Heat ExchangerColumn
5500-1193 (InfinityLab Quick Turn Capillary ST 0.17 mm x 105 mm, long socket)MCT Heat ExchangerColumn
5500-1191 (InfinityLab Quick Turn Capillary ST 0.12 mm x 280 mm, long socket)Column/MCT ValveDetector
5062-8535 (Waste accessory kit (Flow Cell to waste))VWDWaste
5062-2462 (Tube PTFE 0.7 mm x 5 m, 1.6 mm od)DAD/FLDWaste
G5664-68712 (Analytical tubing kit 0.25 mm i.d. PTFE-ESD)DetectorFraction Collector

Capillary and Tubing Connections in Flow Path

p/nFromTo
5500-1596 (Quick Turn Capillary MP35N 0.12 mm x 280 mm)Column/MCT ValveDetector (DAD or 1290 FLD)
5500-1598 (Quick Turn Capillary MP35N 0.12 mm x 500 mm)Column/MCT ValveDetector (VWD)
5062-8535 (Waste accessory kit (Flow Cell to waste))VWDWaste
5062-2462 (Tube PTFE 0.7 mm x 5 m, 1.6 mm od)DAD or 1290 FLDWaste
G5664-68712 (Analytical tubing kit 0.25 mm i.d. PTFE-ESD)DetectorFraction Collector

Figure 38: 1290 Infinity II Bio Online Sample Manager Accessory Kit

#Qty.p/nDescription
1G3167-680101290 Infinity II Bio Online Sample Manager Accessory Kit
12G3167-42000Single Holder UVD Multi Function
225043-1356Column Holder Lamella
312110-1486Fuse 2 AT250 V
415043-0270Leak plane
515043-0271Holder leak plane
615067-4792Leak sensor assembly
715063-6527Tubing, Silicon Rubber, 1.2 m, ID/OD 6 mm/9 mm
815500-1156T-Tube Connector ID6.4
915320-00033-position/6-port FI valve, MP35N, 1300 bar
1015005-0072Capillary MP35N 0.12 mm x 160 mm SL/S
1115005-0069Capillary MP35N 0.17 mm x 160 mm SL/SL
123G4220-60007Bottle Head Assembly
18121-3099CAN Cable, 1 m, flat (not shown)

Figure 42: G3167BA Online Sample Manager Set PM Kit

#Qty.p/nDescription
1G3167-67003G3167BA Online Sample Manager Set PM Kit
11G7137-87201Needle Bio-compatible
21G3167-60017Needle Seat Bio Capillary ST 0.17 x 230 SL/SL (5067-5403 (UHP fitting) is shown as pre-installed but included as a separate part)
325320-0005Rotor Seal, 3-position/6-port FI Valve, MP35N, 1300 bar
41G5668-60494Seal 500 µL Bio-inert

Parts and Consumables — Infinity Lab Quick Connect and Quick Turn Fittings

p/nDescription
5500-1191InfinityLab Quick Turn Capillary ST 0.12 mm x 280 mm, long socket
5500-1192Capillary ST 0.12 mm x 500 mm, long socket
5500-1193InfinityLab Quick Turn Capillary ST 0.17 mm x 105 mm, long socket
5500-1194Capillary ST 0.17 mm x 150 mm, long socket
5500-1234Capillary ST 0.17 mm x 180 mm
5500-1195Capillary ST 0.17 mm x 200 mm, long socket
5500-1196Capillary ST 0.17 mm x 280 mm, long socket
5500-1235Capillary ST 0.17 mm x 380 mm, long socket
5500-1236Capillary ST 0.17 mm x 400 mm, long socket
5500-1197Capillary ST 0.17 mm x 500 mm, long socket
5500-1237Capillary 0.17 mm x 700 mm, ns/ns
5500-1262Capillary 0.25 mm x 150 mm, ns/ns
5500-1263Capillary ST 0.25 mm x 400 mm, long socket
5500-1200Quick Turn Capillary ST 0.12 mm x 130 mm SL/M
5500-1288Capillary ST 0.12 mm x 150 mm, long socket, M4
5500-1290Capillary ST 0.17 mm x 150 mm, long socket, M4

Engineering Facts

14 facts · every claim linked to a page
Selection

The system is intended for automated process monitoring with real-time data, so it suits process analytics rather than routine offline QC.

Confirms the intended application before selecting this platform for a process monitoring task.

Source
“The 1260 Infinity III Prime Online LC System is a process analysis system enabling automated process monitoring, providing real-time data for greater control and faster understanding of processes.”
Selection

The Online Sample Manager supports automated dilutions up to 1:1000, which defines the maximum sample dilution capability.

Determines whether the module can handle concentrated process samples without manual prep.

Source
“The Online Sample Manager provides automated dilutions of up to 1:1000, retain-sample functionality, and direct analysis of the process samples, as well as automated sample preparation.”
Selection

The DAD WR covers 190–950 nm with a 120 Hz acquisition rate, so it fits fast LC and broad UV-Vis detection needs.

Guides detector choice when fast acquisition or wide wavelength range is required.

Source
“With its 120 Hz data acquisition rate the detector is perfectly suited for fast LC applications.”
Selection

The MCT thermostats columns from 20 °C below ambient up to 110 °C, defining the usable column temperature window.

Needed to verify the column oven can reach the required method temperatures.

Source
“facilitates precise column thermostatting over a broad temperature range with cooling down to 20 °C below ambient temperature and heating up to 110 °C.”
Installation

Stack configuration should be chosen to minimize delay volume, so the flow path layout is a performance decision.

Affects how modules are physically arranged to keep delay volume low.

Source
“These configurations optimize the system flow path, ensuring minimum delay volume.”
Installation

A single stack is limited to a maximum of four modules, which constrains how many modules can be stacked together.

Determines whether the planned module count fits a single-stack layout.

Source
“Single Stack (maximal 4 modules, in a bench rack or directly on the bench)”
Installation

Two stacks are recommended when stack height becomes excessive, for example with an additional detector.

Informs the physical layout decision when adding detectors.

Source
“To avoid excessive height of the stack (for example when using the system in combination with an additional detector) it is recommended to form two stacks.”
Safety

The solvent cabinet holds a maximum of 8 L total and no individual bottle may exceed 2 L.

Sets capacity limits for solvent storage during installation.

Source
“The solvent cabinet is designed to store a maximum volume of 8 L solvent. The maximum volume for an individual bottle stored in the solvent cabinet should not exceed 2 L.”
Safety

The leak sensor stops the running system once the leak detection level is reached, so leak handling is automatic.

Explains the automatic shutdown behavior relied on for safe operation.

Source
“The leak sensor in the leak plane stops the running system as soon as the leak detection level is reached.”
Safety

DMF must not be used as mobile phase because it corrodes the PVDF leak sensor.

Prevents solvent selection that would damage the leak sensor.

Source
“Do not use DMF as mobile phase.”
Installation

Waste drainage must run straight into the waste container without restriction at bends or joints.

Guides waste tubing routing to avoid blockages and backflow.

Source
“The waste drainage must go straight into the waste containers. The waste flow must not be restricted at bends or joints.”
Operation

The ERI interface timeout defines how long the software waits for a sample delivery device answer before marking the sample missed.

Needed to configure reliable communication with the sample delivery hardware.

Source
“The Timeout [min] value defines the time, how long the software waits for an answer from the sample delivery device.”
Safety

Solvents with auto-ignition temperature below 200 °C or boiling point below 56 °C must not be used.

Constrains mobile phase selection for safe operation.

Source
“Do not use solvents with an auto-ignition temperature below 200 °C (392 °F). Do not use solvents with a boiling point below 56 °C (133 °F).”
Safety

Bottles exceeding the maximum permissible volume of 2.5 L must not be used.

Sets a bottle size limit for safe solvent handling.

Source
“Do not use bottles that exceed the maximum permissible volume (2.5 L).”