Agilent 4500 Series FTIR Operation Manual

Contenido

  1. Overview pp. 5-6
    Overview of the 4500 series FTIR instruments and their features
  2. Procedure pp. 7-16
    Packing list, power connection, power cord selection, startup, and diagnostics
  3. Procedure pp. 17-38
    Sample analysis and accessory cleaning for TumblIR, ATR, and DialPath
  4. Ordering pp. 37-38
    Spare parts list for the instrument and accessories
  5. Spec pp. 39-41
    Technical specifications and safety information
  6. p. 14 table, 6 rows
  7. p. 37 table, 2 rows
  8. p. 12 table, 8 rows

Table 2. Diagnostic values

ValueOptimal (green)Marginal (yellow)Critical (red)Comment
Energy (centerburst)26,000–21,000>30,000 or <15,000>31,000 or <10,000Indicates the overall alignment and proper gain adjustment for the system.
Battery lifeAC or >30 min30–5 min<5 min
Source1.9 A>2.2 or <1.6>2.5 or <1.0Indicates a problem with the source control voltage or that the source has burnt out. Both voltage and amperage are displayed; however, amperage provides a sufficient diagnostic.
Laser signal15,000–3,000>15,000 or <3,000>17,000 or <2,000Can verify gross alignment errors even if the reflectance cap is not in place.
Detector temperature35–44<35 or >44<30 or >48Indicates a problem with the cooling circuitry or an ambient temperature over the specified range.
CPU temperature10–75>75>80Indicates the ambient temperature over the specified range.

Table

Part numberDescription
430 - 0001Agilent 4500 Series FTIR power supply
430 - 0018Agilent 4500 Series FTIR USB cable

Table 1. Power switch LED status

StatusLED colorAction (% duty cycle)
System OFFNo LED illuminationN/A
System OFFRed100%
System startupRed/greenGreen 0.5 s on / Red 0.5 s on
System ONGreen100%
Low batteryGreen50% (0.5 s on / 0.5 s off)
Critically low batteryRed50% (0.5 s on / 0.5 s off)
Dead batteryNo LED illuminationN/A
Firmware updateRed/greenGreen blinks rapidly twice then red blinks rapidly twice

Engineering Facts

14 facts · every claim linked to a page
Safety

The instrument is not intrinsically safe, so it must only be used in atmospheres verified free of flammable materials.

Determines whether the analyzer can be installed in a classified or potentially explosive area.

Source
“4500 Series FTIR systems are NOT intrinsically safe. Use the system only in atmospheres that have been tested for flammable materials.”
Installation

The shipment must be allowed to equilibrate to room temperature for several hours before opening to prevent condensation on the optics.

Prevents condensation damage to the interferometer and optics during initial setup.

Source
“place the shipment in a room-temperature environment and allow several hours for the contents of the container to reach the room ambient temperature.”
Installation

The instrument must be placed on a flat, stable surface away from hot surfaces and electromagnetic interference sources.

Defines siting constraints that affect measurement stability and interference.

Source
“place it on a flat, stable surface. The instrument must be kept away from hot surfaces and any sources of electromagnetic interference.”
Installation

The power cord must be connected to the spectrometer first and then to the outlet to avoid startup damage.

Wrong connection order can cause power startup damage to the electronics.

Source
“connect the power cord to the spectrometer system first and then to the power outlet.”
Installation

A UL-approved surge protection strip must always be used between the outlet and the system power cord.

Protects the instrument from power surges and faulty power sources.

Source
“always use a UL approved power surge protection strip between the power outlet source and the system power cord.”
Operation

The power button is a momentary switch requiring a two-second hold to turn the system on or off.

Prevents accidental power-off and explains why a short press does nothing.

Source
“In order to turn the system on or off, hold the button down for two (2) seconds.”
Operation

The spectrometer needs a 5-minute warmup before reliable analysis, and 30 minutes before performance validation tests.

Sets the minimum wait time before collecting valid data after power-on.

Source
“The spectrometer must have a warmup period of 5 minutes before reliable analysis may begin.”
Performance

Diagnostic values should be checked daily or weekly, and performance validation quarterly or bi-annually, more often in regulated industries.

Defines the routine verification schedule needed to trust measurement results.

Source
“the diagnostic values should be checked on a daily or weekly basis.”
Maintenance

ZnSe windows are damaged by samples outside pH 4 to 9, and only soft cotton swabs with acetone should be used for cleaning.

Determines which samples are compatible and how to clean without destroying the windows.

Source
“ZnSe is relatively chemically-resistant to materials with a pH range between 4 and 9, but there are some materials, such as strong acids or heavily basic materials, that can harm ZnSe.”
Operation

The TumblIR must not be used with solid samples because they will damage the ZnSe windows or alter the fixed pathlength.

Prevents accessory damage and loss of the factory-set 100 µm pathlength.

Source
“the TumblIR is not to be used with any solid samples, such as hard graphite lubricant. Use of the TumblIR with solid samples will damage the ZnSe windows.”
Selection

The single-reflection ATR suits high-absorbing, powder and solid samples and limited sample amounts, while the triple-reflection ATR suits low-absorbing liquids.

Guides the choice between single- and triple-reflection ATR configurations for the sample type.

Source
“The single reflection ATR is most suitable for higher absorbing samples such as rubber, polymers, paints and fibers.”
Operation

The ATR sample press applies 15 pounds of force assuming a 0.03 in sample thickness and must not be used on sharp or pointed samples.

Defines the contact pressure needed for good spectra and the risk of cracking the thin diamond.

Source
“The spring-loaded sample press tip is set to apply 15 pounds of force assuming a 0.03 in sample thickness.”
Operation

The background must be measured at the same pathlength as the sample on the DialPath.

Prevents invalid results from mismatched background and sample pathlengths.

Source
“The background measurement should be taken at the same pathlength that the sample will be measured at.”
Maintenance

Breaking the spectrometer seal to clean interior surfaces voids the warranty.

Warns service personnel not to open the enclosure for cleaning.

Source
“Do not break the spectrometer seal and attempt to clean interior surfaces. Breaking the seal will void the warranty.”