Agilent Cary Eclipse Fluorescence Spectrophotometer

Contenido

  1. Overview pp. 1-2
    Introduces the Cary Eclipse fluorescence spectrophotometer and its key benefits
  2. Overview p. 3
    Describes fast scanning capability and measurement speed advantages
  3. Spec p. 4
    Covers optical design, sensitivity, and performance specifications
  4. Overview p. 5
    Describes Cary WinFLR software features and usability
  5. Overview p. 6
    Summarizes the range of applications supported by the instrument
  6. Procedure p. 7
    Explains temperature control monitoring and accessory use
  7. Overview pp. 8-9
    Covers chemical and materials applications including detergent fluorescence measurement
  8. Overview pp. 10-11
    Describes life science applications such as molecular rotational motion studies
  9. Services p. 12
    Presents Agilent CrossLab support and service offerings

Engineering Facts

14 facts · every claim linked to a page
Maintenance

The xenon flash lamp has a rated lifetime of 3 billion flashes, typically lasting 10 years, which minimizes replacement cost over the instrument's life.

Long lamp life directly lowers cost of ownership and planned maintenance intervals when selecting the instrument.

Source
“with a lifetime of 3 billion flashes, the lamp typically lasts 10 years”
Performance

The instrument scans at up to 24,000 nm/min, covering the wavelength range in under 3 seconds and collecting 80 points per second for kinetics.

Scan speed and kinetics sampling rate determine whether the instrument meets fast-reaction measurement needs.

Source
“with a scan rate of up to 24,000 nm/min, you can scan the wavelength range in under 3 seconds and collect 80 points per second for kinetics measurements”
Performance

The monochromator grating moves only while the lamp is off, so fast scanning produces no peak shifts.

Confirms that high scan rates do not compromise wavelength accuracy, which matters for reliable spectral data.

Source
“The grating is moved only when the lamp is off, resulting in a go-stop-flash method of taking a measurement.”
Selection

Reduced-volume cuvettes down to 5 µL are supported, with a horizontal beam profile maintaining sensitivity at low volumes.

Defines the minimum sample volume the system can handle, critical for precious or biological samples.

Source
“Reduced volume cuvettes as low as 5 μL are available.”
Installation

The instrument requires only 600 mm of bench space.

Bench footprint is a key constraint when planning lab layout and installation.

Source
“The Cary Eclipse occupies only 600 mm (24 inches) of bench space.”
Operation

The xenon flash lamp and signal processing provide room-light immunity, allowing measurements with the sample compartment lid off.

Enables open-compartment work with large or odd-shaped samples without light shielding.

Source
“The unique xenon flash lamp and sophisticated signal processing give the Cary Eclipse room light immunity for fluorescence measurements.”
Performance

Red-sensitive photomultiplier detectors extend sensitivity up to 900 nm without sacrificing UV performance.

Spectral range coverage determines suitability for red-emitting dyes and NIR-adjacent applications.

Source
“Red-sensitive Photomultiplier tube detectors extend sensitivity up to 900 nm without sacrificing UV performance.”
Operation

The lamp flashes only when a reading is taken, giving zero warmup time, low energy use, and eliminating photodegradation of photosensitive samples.

Zero warmup and low power draw affect daily workflow and operating costs.

Source
“The lamp only flashes when a reading is taken. This means zero warmup time, low electrical energy use, and low maintenance requirements.”
Performance

Signal-to-noise mode reduces scanning times by over 50% by scanning fast in high-emission regions and increasing averaging where emission is low.

Affects throughput and data quality trade-offs for samples with widely varying emission intensity.

Source
“S:N mode reduces scanning times by over 50% as the system scans quickly in areas of high emission intensity and increases signal averaging when the emission is less.”
Selection

The microplate reader accessory converts the instrument into a high-throughput reader in under 30 seconds, measuring 96 wells in under 50 seconds and 384 in under 90 seconds.

Throughput figures are decisive when choosing between cuvette and plate-based workflows.

Source
“Measure 96 wells in less than 50 seconds and 384 in less than 90 seconds.”
Installation

The microplate reader uses reflective optics instead of fiber optics and allows control of the image spot size down to 2 mm diameter.

Spot size and optics type affect measurement alignment and sensitivity for small-volume wells.

Source
“Control the image spot size to as small as 2 mm in diameter.”
Operation

The Peltier thermostatted cell holder controls up to four samples with typical stability of ±0.05 °C and maximum cell-to-cell difference of 0.2 °C at 37 °C.

Temperature stability is essential for reproducible temperature-dependent fluorescence measurements.

Source
“Excellent stability control over time (typical variation ±0.05 °C)”
Operation

Temperature ramp rates as slow as 0.06 °C/min can be selected for thermal denaturation and renaturation studies via FRET.

Slow controlled ramps are required for accurate thermal melting experiments.

Source
“Temperature ramp rates as slow as 0.06 °C/min can be selected for thermal denaturation and renaturation studies of DNA via fluorescence resonance energy transfer (FRET)”
Selection

Optional UV transmitting film polarizers enable excitation down to 275 nm, allowing tryptophan excitation without photobleaching.

Defines the UV polarization capability needed for protein rotational and anisotropy studies.

Source
“Optional UV transmitting film polarizers enable excitation down to 275 nm, ensuring that even tryptophan can be excited without photobleaching.”