US EPA Method 624 with the Tekmar Lumin and the Agilent 7890B
Technical Reference Summary
Covered Models & Document Scope
Nota de aplicación para el concentrador de purga y trampa Teledyne Tekmar Lumin con automuestreador AQUATek 100, utilizado con un GC/MS Agilent 7890B para el análisis de COV según el Método 624 de la US EPA; abarca las condiciones de operación del Lumin/AQUATek 100 y las condiciones del GC/MS, no es una hoja de datos para pedidos de producto.
Technical Data
| Parameter | Value | Condición |
|---|---|---|
| Valve oven temperature | 150 °C p.2 ↗ | Lumin, standby |
| Transfer line temperature | 160 °C p.2 ↗ | Lumin, standby |
| Sample mount temperature | 90 °C p.2 ↗ | Lumin, standby |
| Standby flow | 20 mL/min p.2 ↗ | Lumin, standby |
| Purge ready temperature | 35 °C p.2 ↗ | Lumin, standby |
| MCS purge temperature | 20 °C p.2 ↗ | Lumin, purge |
| Purge temperature | 20 °C p.2 ↗ | Lumin, purge |
| Purge time | 11.00 min p.2 ↗ | Lumin, purge |
| Purge flow | 40 mL/min p.2 ↗ | Lumin, purge |
| Dry purge temperature | 20 °C p.2 ↗ | Lumin, purge |
| Dry purge time | No Dry Purge, 0 min p.2 ↗ | Lumin, purge |
| Desorb preheat temperature | 245 °C p.2 ↗ | Lumin, desorb |
| Desorb temperature | 250 °C p.2 ↗ | Lumin, desorb |
| Desorb time | 1.00 min p.2 ↗ | Lumin, desorb |
| Drain flow | 300 mL/min p.2 ↗ | Lumin, desorb |
| Bake time | 2.00 min p.2 ↗ | Lumin, bake |
| Bake temperature | 280 °C p.2 ↗ | Lumin, bake |
| MCS bake temperature | 180 °C p.2 ↗ | Lumin, bake |
| Bake flow | 200 mL/min p.2 ↗ | Lumin, bake |
| Sample loop time | 0.35 min p.2 ↗ | AQUATek 100 |
| Sample transfer time | 0.35 min p.2 ↗ | AQUATek 100 |
| Rinse loop time | 0.30 min p.2 ↗ | AQUATek 100 |
| Sweep needle time | 0.30 min p.2 ↗ | AQUATek 100 |
| Presweep time | 0.25 min p.2 ↗ | AQUATek 100 |
Operating Limits
- Calibration range: 0.5 to 200 ppb p.2 ↗
- Trap used: #9 p.2 ↗
- Water temperature: 90 °C p.2 ↗
- Bake rinse drain cycles: 2 p.2 ↗
- Bake rinse drain time: 0.35 min p.2 ↗