Fittings, capillaries, solvent bottles, and bench accessories
p. 31table, 4 rows
p. 37table, 24 rows
p. 67table, 2 rows
Table 2: Capillary properties
Color
Inner diameter
µL per cm
Red
0.12 mm (0.005 inch)
0.113
Green
0.17 mm (0.007 inch)
0.227
Blue
0.25 mm (0.01 inch)
0.491
Orange
0.50 mm (0.02 inch)
Table 3: Compatibility of drain connectors and modules
Drain Connector Type
Compatible Module
Compatible Module Type
Double
G7116A/B
Column Compartment
Single
G7114A/B
Detector
G7115A
G7117A/B/C
G7121A/B
G7123B
G7162A/B
G7165A
G7129A/B/C
Sampler
G7167A/B/C
G5668A
G7137A/B
G7157A
G4767A
G7122A
Degasser
G7104A/C
Pump
G7110B
G7111A/B
G7112B
G7120A
G7131A/C
G7132A
G5654A
G4782A
Table
p/n
Description
5067-6168
InfinityLab Quick Connect Assy ST 0.17 mm x 220 mm
5067-6169
InfinityLab Quick Connect Assy ST 0.17 mm x 280 mm
Engineering Facts
14 facts · every claim linked to a page
Selection
The binary pump supports ultrahigh pressure up to 1300 bar and analytical flow rates up to 5 mL/min, so it fits narrow-bore and sub-2 µm particle UHPLC methods.
Defines the pressure and flow envelope that must match the column and method requirements at selection.
Source
“combines ultrahigh pressure up to 1300 bar and high analytical flow rates up to 5 mL/min”
Performance
Low delay volumes down to 10 µl enable fast gradients across the full 1–99% composition range.
Delay volume drives gradient speed and resolution, so it matters when configuring the flow path.
Source
“Exceptionally low delay volumes down to 10 µl and high flow rates enable fast gradients over the entire composition range from 1 to 99%”
Selection
The Vialsampler can include an integrated column compartment for two columns up to 80 °C and a sample thermostat from 4 °C to 40 °C in one module.
Determines whether temperature-controlled separations and sample cooling can be met without extra modules.
Source
“temperature control up to 80 °C as well as a sample thermostat for stable temperatures from 4 °C to 40 °C”
Selection
The VWD offers data rates up to 240 Hz with low noise and drift, which matters for fast and trace-level detection.
Data rate and noise determine suitability for fast UHPLC peaks and trace quantification.
Source
“High productivity can be achieved with fast analysis at up to 240 Hz data rates.”
Installation
Stacking modules in the prescribed single-stack order optimizes the flow path for minimum delay volume and minimizes bench space.
Correct stack order directly affects delay volume and available bench space during installation.
Source
“This configuration optimizes the flow path for minimum delay volume and minimizes the bench space required.”
Installation
When an additional detector makes the stack too tall, the system should be split into two stacks.
Prevents unsafe or unstable tall stacks when extra detectors are added.
Source
“To avoid excessive height of the stack (for example when using the system in combination with an additional detector), it is recommended to form two stacks.”
Installation
The solvent cabinet holds a maximum of 8 L total, and no individual bottle may exceed 2 L.
Limits solvent bottle sizing and placement during system setup.
Source
“The solvent cabinet is designed to store a maximum volume of 8 L solvent. The maximum volume for an individual bottle stored in the solvent cabinet should not exceed 2 L.”
Safety
The leak sensor in the leak plane stops the running system once the leak detection level is reached.
Explains the automatic safety shutdown behavior that must be understood before operation.
Source
“The leak sensor in the leak plane stops the running system as soon as the leak detection level is reached.”
Installation
Drain connectors must be installed on every module except the lowest one, and the lowest module must use a waste tube instead.
Incorrect drain connector placement causes leaks or poor drainage of viscous solvents.
Source
“Make sure that dripping adapters are correctly installed on each module in the LC stack, excluding lowest module.”
Installation
Drain connectors must not be installed on the bottom module; its drain outlet must go to a waste container via tubing.
Prevents improper drainage configuration at the base of the stack.
Source
“Do not install drain connectors on the bottom modules of the stack.”
Safety
If a custom waste container is used, the waste tubes must not be immersed in the liquid.
Prevents siphoning and back-pressure issues in the waste path.
Source
“If you decide to use your own waste solution, make sure that the tubes don't immerse in the liquid.”
Installation
LAN configuration must be done from the module with the direct LAN connection, usually the detector with the highest data rate.
Determines which module must be cabled to the network for reliable control.
Source
“LAN configuration is executed from the module with direct LAN connection to the controller software. This must be the module (usually the detector) with the highest data rate.”
Maintenance
Each channel should be purged with fresh solvent at 2.5–3 mL/min for 5 minutes before operation.
Defines the daily purge routine needed for stable baselines and correct operation.
Source
“Purge each channel with fresh solvent at 2.5 – 3 mL/min for 5 min before operation.”
Maintenance
The inline filter should be back-flushed weekly and its frit exchanged regularly to protect the downstream system.
Establishes the maintenance interval that prevents blockages and precipitation issues.