NovoSampler Pro Automated Sampling System User Manual
Sommaire
pp. 3-4
Document contents and section listing
Warningpp. 5-9
Safety notices, symbols, and technical support contacts
Overviewpp. 10-13
System overview, components, and plate/tube identification
Procedurepp. 14-16
Mounting sampler to flow cytometer and calibration steps
Overviewpp. 17-34
Software interface, settings, and SIP movement offset
Procedurepp. 35-47
Plate selection, active sample setup, and acquisition workflow
Procedurep. 48
Cleaning procedures for NovoSampler Pro
pp. 49-55
Error codes, symptoms, and corrective actions
pp. 56-57
Supplementary reference information and tables
p. 50table, 15 rows
p. 54table, 7 rows
p. 52table, 10 rows
p. 53table, 13 rows
p. 28table, 6 rows
p. 57table, 14 rows
p. 55table, 4 rows
p. 51table, 9 rows
p. 27table, 6 rows
p. 3table, 40 rows
Troubleshooting
Message IDs
Software Messages
Possible Causes
Recommended Solutions
0x0001
Collision of SIP
Dirty SIP or SIP wash apparatus
Clean the SIP or SIP wash apparatus following the procedures described in Section 1.1 Preventative Maintenance in NovoCyte® Flow Cytometer Maintenance Guide.
0x0002
Running out of NovoFlow
NovoFlow solution is not sufficient to continue to run any samples
Refill the NovoFlow container following the procedures described in Section 2.1.2 Add Instrument Reagents in NovoCyte® Flow Cytometer Operator’s Guide.
0x0002
NovoCyte Fluidics Station is not working properly
Replace the NovoCyte Fluidics Station.
0x0003
Running out of NovoRinse
NovoRinse solution is not sufficient to continue to run any samples
Refill the NovoRinse container following the procedures described in Section 2.1.2 Add Instrument Reagents in NovoCyte® Flow Cytometer Operator’s Guide.
0x0003
NovoCyte Fluidics Station is not working properly
Replace the NovoCyte Fluidics Station.
0x0004
Running out of NovoClean
NovoClean solution is not sufficient to continue to run any samples
Refill the NovoClean container following the procedures described in Section 2.1.2 Add Instrument Reagents in NovoCyte® Flow Cytometer Operator’s Guide.
0x0004
NovoCyte Fluidics Station is not working properly
Replace the NovoCyte Fluidics Station.
0x0005
Waste container is full
Waste container is too full to continue to run any samples
Empty the waste container following the procedures described in Section 2.1.3 Empty Waste in NovoCyte® Flow Cytometer Operator’s Guide.
0x0005
NovoCyte Fluidics Station is not working properly
Replace the NovoCyte Fluidics Station.
0x0006
System voltage is out of range
System Error
►►Restart the NovoCyte instrument. ►►Contact ACEA technical support if the error persists.
0x0007
System electric current is out of range
System Error
►►Restart the NovoCyte instrument. ►►Contact ACEA technical support if the error persists.
0x0008
Firmware configuration error
System Error
►►Restart the NovoCyte instrument. ►►Contact ACEA technical support if the error persists.
0x0009
0x000A
0x000B
…… laser self test error
Specified laser is not functioning properly
The instrument will automatically reset the laser and run a laser self-test. It takes approximately 5 to 10 minutes. ►►►
Troubleshooting
Message IDs
Software Messages
Possible Causes
Recommended Solutions
0x0106
NovoSampler (Pro) is disconnected when powered up
The NovoSampler (Pro) is disconnected when powered up
►Shut down NovoCyte Flow Cytometer. ►Reconnect the cable of the NovoSampler (Pro) to NovoCyte flow cytometer. ►Turn on NovoCyte and follow the prompts to calibrate the NovoSampler (Pro). ►Contact ACEA technical support if the error persists.
0x0109
NovoCyte Fluidics Station is not connected
Cable between the NovoCyte Fluidics Station and the NovoCyte instrument is not properly connected
►Turn off the instrument. ►Reconnect the cable between the NovoCyte Fluidics Station and the NovoCyte instrument. ►Power up the instrument. ►Contact ACEA technical support if the error persists.
0x010A 0x010B 0x010C 0x010D
… liquid level sensor failure
Specified liquid level sensor in the NovoCyte Fluidics Station or NovoCyte Fluidics Cart is not working properly
►Reconnect the NovoCyte Fluidics Station or NovoCyte Fluidics Cart cable. ►Restart the NovoCyte instrument. ►Replace the NovoCyte Fluidics Station or NovoCyte Fluidics Cart. ►Contact ACEA technical support if the error persists.
0x010F
Sheath filter is clogged. Please replace the sheath filter and run the Priming procedure
Sheath in-line filter is clogged
Replace the sheath in-line filter following the procedures described in Section 1.2 Replacing Fluidic System Consumables in NovoCyte® Flow Cytometer Maintenance Guide.
0x0110
Recovering collision error
Plate stops at an incorrect position
No action is needed. Instrument will automatically recover the error.
0x0111
Orbital shaker homing position reset failure
Orbital shaker of NovoSampler Pro is not working properly
No action is needed. The NovoSampler Pro can be used normally.
0x7000
NovoCyte Fluidics Cart firmware error
NovoCyte Fluidics Cart firmware is not working properly
►Restart the instrument ►Replace the NovoCyte Fluidics Cart ►Contact ACEA technical support if the error persists.
Troubleshooting
Message IDs
Software Messages
Possible Causes
Recommended Solutions
0x0014
Pressure is out of limit
Waste container is not correctly connected to the instrument
Check the quick coupling connectors to ensure the waste container is correctly connected to the instrument.
0x0014
Pressure is out of limit
Sheath in-line filter is clogged
Replace the sheath in-line filter following the procedures described in Section 1.2 Replacing Fluidic System Consumables in NovoCyte® Flow Cytometer Maintenance Guide.
0x0014
Pressure is out of limit
Flow cell is clogged
Conduct Unclog from NovoExpress.
0x0015
NovoSampler firmware error
NovoSampler or NovoSampler Pro firmware is not working properly
Re-install or upgrade the NovoSampler or NovoSampler Pro firmware.
0x0016 0x0017 0x0018
…… laser does not emit
Specified laser is not connected properly or laser is not working properly
►Reconnect the laser cable properly. ►Restart the instrument. ►Contact ACEA technical support if the error persists.
0x0019 0x001A 0x001B
…… laser communication error
Interference or bad connection or bad laser
►Reconnect the laser properly. ►Restart the instrument. ►Contact ACEA technical support if the error persists.
0x001C
Sample injection probe reset failed
Bad connection or optocoupler damaged
►Click the OK button in the prompted dialog box or wait 10 seconds for automatically error handling. ►Restart the instrument. ►Contact ACEA technical support if the error persists.
0x001D
Sampling Pump reset failed
Bad connection or optocoupler damaged
►Restart the instrument. ►Contact ACEA technical support if the error persists.
0x001E
The SIP module needs to be upgraded to support NovoSampler Pro
SIP module is not compatible with connected NovoSampler Pro
Contact ACEA technical support for service.
0x0020
System initialization is paused
Liquid level in the reagent containers is not within normal range when NovoCyte is powered up.
►Make sure that the instrument reagent containers are placed correctly and the liquid level in the containers is within the normal range. ►Click OK on the prompted dialog to continue system initialization.
Troubleshooting
Message IDs
Software Messages
Possible Causes
Recommended Solutions
0x0021
Sheath fluid pump reset failed
Bad connection or optocoupler damaged
► Restart the instrument. ► Contact ACEA technical support if the error persists.
0x0022
Resetting NovoSampler Pro to zero position failed
NovoSampler Pro is not working properly.
Click Reset. Restart the NovoSampler Pro.
0x0023
Resetting NovoSampler Pro to home position failed
NovoSampler Pro is not working properly.
Click Reset. Restart the NovoSampler Pro.
0x0100
Instrument cover opened
Instrument cover is open or not tightly closed
Close the instrument cover.
0x0101
NovoFlow running low
NovoFlow solution is below the pre-set volume limit
Refill the NovoFlow container following the procedures described in Section 2.1.2 Add Instrument Reagents in NovoCyte® Flow Cytometer Operator’s Guide.
0x0101
NovoFlow running low
NovoCyte Fluidics Station not working properly
Replace NovoCyte Fluidics Station.
0x0102
NovoRinse running low
NovoRinse solution is below the pre-set volume limit
Refill the NovoRinse container following the procedures described in Section 2.1.2 Add Instrument Reagents in this guide.
0x0102
NovoRinse running low
NovoCyte Fluidics Station not working properly
Replace NovoCyte Fluidics Station.
0x0103
NovoClean running low
NovoClean solution is below the pre-set volume limit
Refill the NovoClean container following the procedures described in Section 2.1.2 Add Instrument Reagents in NovoCyte® Flow Cytometer Operator’s Guide.
0x0103
NovoClean running low
NovoCyte Fluidic Station is not working properly
Replace NovoCyte Fluidics Station.
0x0104
Waste container is close to full
Waste is above the pre-set volume limit
Empty the waste container following the procedures described in Section 2.1.3 Empty Waste in NovoCyte® Flow Cytometer Operator’s Guide.
0x0104
Waste container is close to full
NovoCyte Fluidics Station not working properly
Replace NovoCyte Fluidics Station.
0x0105
Cover of NovoSampler (Pro) opened
Cover of NovoSampler (Pro) is opened
Close the cover.
Plate Manager Panel
Icon
Description
Mix: Sets the mixing conditions. Mix can be set for None (no mixing), 1 Cycle, 2 Cycles, 3 Cycles for every defined number of wells or every defined number of minutes. Default mixing conditions (i.e. Mix 1 Cycle Every 1 Well) will be automatically applied and cannot be modified by user when Standard or HT mode is selected.
Rinse: Sets the rinse cycle to clean the sample injection probe. Rinse can be set for None (no rinse), 1 Cycle, 2 Cycles, 3 Cycles for every defined number of wells. Default rinse conditions (i.e. Rinse 1 Cycle Every 1 Well when Standard mode is selected, or None when HT mode is selected) will be automatically applied and cannot be modified by user when Standard or HT mode is selected. There is always one rinse cycle at the end of each plate run even the Rinse setting is set to None.
Mixing Parameters: Sets the mixing parameters for the orbital shaker. When NovoSampler Pro works in Custom mode, these mixing parameters for each plate type are defined in the Manage Plate Type window from the Instrument → NovoSampler Pro menu, and can be modified in this area for the current experiment. When NovoSampler Pro works in Standard or HT mode, default mixing parameters will be automatically applied and cannot be modified by user. Please refer to Table 2.1 in this guide for more details.
Speed: The mix speed of the orbital shaker.
Acceleration: The acceleration time for the orbital shaker to reach the specified Mix Speed.
Duration: The mixing time for the orbital shaker to mix the sample in one cycle.
Technical Specifications
Parameter
Specification
NovoSampler Pro Dimension (W × D × H)
14.6 × 10.2 × 11.0 in (37 × 26 × 28 cm)
NovoSampler Pro Weight
19.8 lb (9 kg)
NovoCyte Flow Cytometer with NovoSampler Pro installed: Work space requirements (W × D × H)
186 × 80 × 80 cm
Operating temperature
15 °C ~ 32 °C
Operating Humidity
< 80% and non-condensing
Power requirements
100 ~ 220 VAC, 47 ~ 63 Hz, ≤ 20 W
Compatibility
Compatible with standard 96-well plates (flat-bottom, U-bottom, V-bottom), standard 24-well, 48-well plates, and NovoSampler 24 tube rack for 12 × 75 mm tubes
Carryover rate
less than 0.1% (with one rinse cycle per well)
Mix mode
orbital shaker
Rinse
automatically rinse the sample injection probe and fluidic tubing.
Minimum sample volume
90 µL for 96-well plates
Dead volume
50 µL for 96-well plates
Operating System
Microsoft Windows®
Software
NovoExpress
Table 5-2 NovoExpress Prompted Error Messages and Message IDs
Message IDs
Software Messages
Possible Causes
Recommended Solutions
0x7001
NovoFlow liquid level sensor is not connected
NovoFlow liquid level sensor is not connected or does not work properly
►Reconnect the NovoFlow liquid level sensor
►Contact ACEA technical support if the error remains after reconnection. The sensor may be damaged and need to be replaced.
0x7002
Waste liquid level sensor is not connected
Waste liquid level sensor is not connected or does not work properly
►Reconnect the NovoFlow liquid level sensor
►Contact ACEA technical support if the error remains after reconnection. The sensor may be damaged and need to be replaced.
0x7100
NovoCyte Fluidics Cart communication lost
NovoCyte Fluidics Cart is not properly connected to the instrument
►Ensure the DVI cable between NovoCyte Fluidics Cart and the instrument is properly connected
►Restart the instrument
►Contact ACEA technical support if the error persists.
Communication error (code: xx, xx). Please restart NovoCyte and NovoExpress.
USB cable between the NovoCyte instrument and workstation is not connected properly
►Reconnect the USB cable between the NovoCyte instrument and the workstation.
►Restart the instrument and NovoExpress software.
►Contact ACEA technical support if the error persists.
Troubleshooting
Message IDs
Software Messages
Possible Causes
Recommended Solutions
0x000C
0x000D
0x000E
….laser is not connected
Specified laser is not detected
►Restart the NovoCyte instrument.
►Contact ACEA technical support if the error persists.
0x000F
NovoSampler communication lost
The cable between the NovoSampler (Pro) and NovoCyte Flow Cytometer is not securely connected
Reconnect the cable between the NovoSampler (Pro) and NovoCyte Flow Cytometer.
0x000F
NovoSampler communication lost
NovoSampler (Pro) is not communicating with NovoCyte
►Restart the NovoCyte instrument.
►Contact ACEA technical support if the error persists.
0x0010
NovoSampler (Pro) has not been calibrated
NovoSampler (Pro) is newly installed or re-connected.
Follow the prompted instructions to calibrate the NovoSampler (Pro).
0x0011
NovoSampler (Pro) calibration failed……
NovoSampler (Pro) is not installed properly
Re-install and calibrate the NovoSampler (Pro) following the procedures described in Section 1.2 Installation in NovoSampler® Operator’s Guide or NovoSampler® Pro Operator’s Guide.
0x0011
NovoSampler (Pro) calibration failed……
The NovoSampler (Pro) cover was opened during calibration.
Close the NovoSampler (Pro) cover and redo the calibration.
0x0012
The movement of plate is out of range
The movement of the orbital shaker is blocked.
►Check the path of the orbital shaker to make sure there are no objects blocking the movement. Clear the block if there is any.
►Click Instrument → NovoSampler Pro → Calibrate in NovoExpress Software to re-calibrate the NovoSampler (Pro).
0x0012
The movement of plate is out of range
NovoSampler (Pro) is not installed properly
Re-install and calibrate the NovoSampler (Pro) following the procedures described in Section 1.2 Installation in NovoSampler® Operator’s Guide or NovoSampler® Pro Operator’s Guide.
0x0013
Cover of NovoSampler (Pro) is opened during moving plate
The cover of NovoSampler (Pro) is open during moving plate
Close the NovoSampler (Pro) cover. NovoSampler (Pro) will automatically reset and be ready for operation.
Plate Manager Panel
Icon
Description
Duplicate Sample(s) on Selected Well(s): Creates new samples on empty wells, duplicating the template of the selected samples of the selected wells. If part of the selected wells have samples defined and the rest part have no samples defined, or the selected samples of the selected wells are in different specimen, this tool will be disabled. If the specimen contains more than one sample, and all the samples are selected, clicking this icon will create duplicated samples under a newly created specimen.
New Plate: Creates new plate and specimen, and creates new samples in the new specimen, duplicating the template of the selected samples of the selected wells. The new samples have the same well ID with the selected wells.
Show Run Order: When selected, a blue outline shows on the icon . The sample acquisition sequence will be shown in the plate view.
Mode: User is allowed to select the working mode (Standard, High- Throughput and Custom) of the NovoSampler Pro in this area. Default mixing parameters (i.e. Mixing Speed, Acceleration, Duration, Mix and Rinse pattern), Stop condition and Flow Rate setting will be automatically applied when Standard or HT mode is selected. User can only customize these settings when Custom mode is selected. Selecting the working mode of NovoSampler Pro is only available before the sample is created.
This area is designed to input plate information. The left drop-down box is the Plate ID. The Plate ID is a positive integer and can be modified by highlighting the area and entering a new number. The middle drop-down box is the Plate Type. The drop-down box lists all the available plate type defined in the Manage Plate Types window from the Instrument → NovoSampler Pro menu. Pull down the list and select the corresponding plate type for the experiment. The right box is the Plate Name. Input plate name in this box to help identify the plate. The plate type cannot be modified once the sample acquisition has begun.
This area is designed to input the specimen and sample information. The left drop-down box is the Specimen ID and Specimen Name, separated by a colon. The right box is the Sample Name. When creating new samples, the samples will be created in a specimen with specimen name specified in the Specimen Name box. If the Specimen Name box is empty or does not exist, a new specimen will be created. The default new specimen name is Specimen1, Specimen2,… The default sample name is Sample1, Sample2,…
Table of Content
Раздел
Название
Страница
Prologue
5
I.
Revision History
5
II.
Contact Information
5
III.
Trademark
6
IV.
Intended Use
6
V.
Using this Guide
6
VI.
Conventions
6
VII.
Safety and Limitations
7
VIII.
Technical Support
9
1.
Introduction
10
1.1
Hardware Overview
10
1.2
Installation
13
2.
Software Overview
17
2.1
Software User Interface
17
2.2
Menu Bar
18
2.3
Cytometer Control Panel
22
2.4
Experiment Manager Panel
23
2.5
Plate Manager Panel
24
2.6
Work List
31
2.7
Statistical Tables
33
2.8
Options
34
3.
Sample Acquisition
35
3.1
Preparation
35
3.2
Experiment Setup
36
3.3
Instrument QC Test
37
3.3.1
Prepare QC Particles
37
3.3.2
Run the QC Test
37
3.4
Single Sample Acquisition
38
3.5
Edit Work List
40
3.5.1
Creating Compensation Control Samples
40
3.5.2
Creating Experiment Samples
40
3.5.3
Plate View
43
3.6
Automatic Sample Acquisition
43
3.7
Analysis of Data
46
3.8
Shutdown
47
4.
Cleaning NovoSampler Pro
48
5.
Troubleshooting
49
Appendix
56
Appendix A
Accessories and Consumables
56
Appendix B
Technical Specifications
57
Engineering Facts
14 facts · every claim linked to a page
Selection
Only the NovoSampler 24 tube rack is compatible with the NovoSampler Pro; other rack types can cause malfunction and damage.
Prevents selecting an incompatible rack that would damage the instrument.
Source
“Only NovoSampler 24 tube rack could be used with the NovoSampler Pro.”
Installation
The NovoSampler Pro must be flush against the NovoCyte on the left and back sides; gaps cause calibration failure or probe collision.
Ensures correct alignment and avoids probe collision during installation.
Source
“Make sure there is no gap between the NovoSampler Pro and NovoCyte on both left and back sides.”
Installation
Tighten the mounting screws only finger-tight; excessive force damages the instrument.
Prevents thread or housing damage during mounting.
Source
“Use appropriate force (i.e., finger tight) when tightening the screws.”
Installation
Calibration must be performed successfully before any sample collection is allowed.
Defines the mandatory commissioning step before first use.
Source
“Calibration can be postponed, but the software will not allow for samples to be collected until a calibration is performed successfully.”
Installation
Remove all plates from the orbital shaker before calibration to avoid collision and instrument damage.
Prevents probe collision during the calibration routine.
Source
“Before calibrating the NovoSampler Pro remove any plates in the orbital shaker.”
Operation
The volume stop condition must be set when using the NovoSampler Pro.
Required acquisition setting for correct sampler operation.
Source
“The volume stop condition must be set when using NovoSampler Pro.”
Operation
Run Single Well does not perform mixing automatically; the user must click Mix Sample if mixing is needed.
Affects sample suspension and data quality in single-well mode.
Source
“When Run Single Well is selected, the NovoSampler Pro does not perform the mixing function before sample acquisition.”
Operation
The plate type cannot be changed once sample acquisition has started.
Requires correct plate type selection before starting a run.
Source
“The plate type cannot be modified once the sample acquisition has begun.”
Operation
Enabling the plate-type reminder is recommended to avoid probe collision from a wrong plate selection.
Reduces risk of probe collision and instrument damage.
Source
“It is highly recommended to enable this setting to remind user to check and make sure the correct plate type is selected for the experiment to avoid collision of the sample injection probe.”
Safety
The cover must remain closed during sample acquisition; the lid sensor is a backup, not a substitute for confirming the instrument has stopped.
Protects the operator from moving parts during operation.
Source
“ensure that the cover to the NovoSampler Pro is closed during sample acquisition.”
Safety
Do not reach into the NovoSampler Pro while it is running; the sample injection probe is a moving part that can cause injury.
Prevents operator injury from the moving probe.
Source
“The sample injection probe is a moving part and may cause injury.”
Maintenance
The NovoSampler Pro should only be uninstalled when powered off; uninstalling while powered up may damage it.
Prevents hardware damage during removal or service.
Source
“NovoSampler Pro should only be uninstalled when powered off.”
Performance
The automatic rinse function yields a carryover rate of less than 0.1%.
Quantifies cross-contamination performance for assay design.
Source
“the NovoSampler has a carryover rate of less than 0.1%.”
Performance
Prepared QC particles must be used within four hours after preparation.
Sets the time limit for valid QC testing.
Source
“The prepared QC Particles sample should be used within four hours after preparation.”