Induztr
Résumé technique Visionneuse de documents Modèles associés

Not for use in diagnostic procedures

Sommaire

  1. Spec p. 2
    Lists kit components and quantities
  2. Warning p. 3
    Safety warnings and precautions
  3. p. 1 table, 5 rows
  4. p. 2 table, 5 rows

Kit Specifications

SpecificationsDescription
DNA sizing range100 bp – 6,000 bp
Separation resolution8 bp @ 500 bp (33-55 Array)
DNA sizing precision12% CV
DNA sizing accuracy1± 5% or better
DNA fragment concentration range0.005 ng/μL – 2 ng/μL input DNA (adjustable by dilution of sample)

1 Results using DNA ladder or DNA fragment standards initially prepared with 1x TE buffer

Prepare gel/dye mixture for 5300 (48-Capillary) Fragment Analyzer system:

# of samples to be analyzed1Volume of Intercalating DyeVolume of gel
482.5 µL25 mL
964.0 µL40 mL
1445.5 µL55 mL
1927.0 µL70 mL
2408.5 µL85 mL

1 Typically one sample well per separation is dedicated to the ladder.

Engineering Facts

10 facts · every claim linked to a page
Selection

The kit only supports DNA sizing from 100 bp to 6,000 bp, so samples outside this range cannot be sized with this assay.

Determines whether the kit is suitable for the expected fragment size range before purchase.

Source
“DNA sizing range 100 bp – 6,000 bp”
Performance

Separation resolution is 8 bp at 500 bp with the 33-55 array, which sets the minimum detectable size difference near 500 bp.

Needed to judge if closely sized fragments can be resolved for the application.

Source
“8 bp @ 500 bp (33-55 Array)”
Performance

Input DNA must be between 0.005 ng/μL and 2 ng/μL, adjustable by dilution, so samples must be quantified and diluted before loading.

Prevents overloading or underloading the capillary array and ensures valid sizing.

Source
“0.005 ng/μL – 2 ng/μL input DNA (adjustable by dilution of sample)”
Installation

The gel/dye mixture volume scales with sample count (e.g., 25 mL gel with 2.5 µL dye for 48 samples), so reagent volumes must be planned per run.

Allows correct preparation of fresh gel/dye mixture for the planned number of samples.

Source
“2.5 µL 25 mL”
Maintenance

The 1x 930 dsDNA Inlet Buffer must be replaced daily, which imposes a daily maintenance step on the instrument.

Affects daily operating schedule and buffer consumption planning.

Source
“Replace 1x 930 dsDNA Inlet Buffer daily.”
Maintenance

Capillary Storage Solution must be replaced every 2-4 weeks, setting a recurring maintenance interval.

Defines the long-term capillary storage maintenance cadence.

Source
“Replace Capillary Storage Solution every 2-4 weeks.”
Operation

The MDK DNA Ladder is ready to use without dilution and must be placed at 24 µL into well D12 or H12.

Ensures the ladder is loaded correctly for sizing calibration.

Source
“Place 24 µL of MDK DNA Ladder (“ready to use”; no dilution) into well D12 or H12.”
Operation

The correct software method to select is CRP-910-(33 or 55) – CRISPR Discovery, matching the array type.

Using the wrong method would produce invalid or failed separations.

Source
“select the CRP-910-(33 or 55) – CRISPR Discovery method from the dropdown menu.”
Safety

Reagent handling carries health risks, so safety goggles, gloves, and protective clothing must be worn.

Mandatory PPE requirement for anyone handling kit reagents.

Source
“Follow the appropriate safety procedures such as wearing goggles, safety gloves and protective clothing”
Conformité

The kit is for research use only and must not be used in diagnostic procedures.

Restricts the intended use and prevents clinical/diagnostic application.