The plasma must never be operated unless the torch compartment door is closed with the locking lever fully latched and the space above the chimney is clear.
Interlock and clearance conditions are mandatory for safe plasma ignition and must be verified before operation.
Source
“The plasma must never be operated unless: the torch compartment door is closed, with the locking lever fully latched; and the space above the chimney is clear of objects.”
Safety
The torch and its surroundings remain hot for up to five minutes after plasma extinction, so cooling or heat-resistant gloves are required before work in that area.
Prevents burn injuries during maintenance or torch access immediately after plasma shutdown.
Source
“The torch and its surroundings remain hot for up to five minutes after the plasma is extinguished.”
Installation
The exhaust fan must always be switched on before igniting the plasma to extract heat, ozone, vapors and fumes.
Exhaust must be operational before plasma ignition to avoid hazardous fume exposure.
Source
“The exhaust fan must always be switched on before igniting the plasma.”
Installation
Only instrument-grade argon (99.996% purity) with specified impurity limits is acceptable for the plasma, nebulizer and optics interface purge.
Gas purity directly affects plasma stability, analytical performance and instrument safety.
Source
“Use only ‘instrument grade’ gases with your spectrometer.”
Installation
The spectrometer weighs approximately 203 kg and must be moved with a forklift or other suitable mechanical lifting device.
Site planning and rigging must account for the heavy instrument to avoid injury or damage during installation.
Source
“The spectrometer weighs approximately 203 kg (448 lb).”
Performance
For optimum analytical performance, laboratory ambient temperature should be 20–25 °C and held constant within ±2 °C throughout the working day.
Stable temperature control is required for reproducible optical emission measurements.
Source
“For optimum analytical performance, it is recommended that the ambient temperature of the laboratory be between 20 and 25 °C (68 and 77 °F).and be held constant to within ±2 °C (±3.6 °F) throughout the entire working day.”
Installation
Use of power boards or extension cables is not recommended for the instrument electrical supply.
Direct mains connection avoids voltage drop and electrical safety issues during installation.
Source
“Use of power boards or extension cables is not recommended.”
Installation
The water flow sensor stops plasma and camera Peltier operation if cooling water flow drops below 1.1 L/min.
Cooling water flow must be maintained above the trip threshold to prevent instrument shutdown and damage.
Source
“The instrument is equipped with a water flow sensor, which will stop operation of the plasma and camera Peltier assembly if the cooling water flow through the instrument drops below 1.1 L/min (0.3 gpm).”
Installation
The water cooling system must always be on before igniting the plasma.
Cooling must be established before plasma ignition to protect the torch and instrument.
Source
“Always ensure the water cooling system is on before igniting the plasma.”
Installation
A chemically inert drain vessel of at least 2 liters, not glass or narrow-necked, must be provided by the user and located under the sample compartment or on the right side of the instrument.
Proper waste drainage capacity and material are required for safe continuous operation.
Source
“A chemically inert container, not glass or of a narrow-necked style, to hold a minimum of 2 liters (4 pints) of waste must be provided by the user.”
Installation
The Agilent 700 Series ICP-OES must be installed by an Agilent-trained, Agilent-qualified or Agilent-authorized field service engineer.
Installation by authorized personnel is a prerequisite for warranty and safe operation.
Source
“The Agilent 700 Series ICP-OES must be installed by an Agilent-trained, Agilent-qualified or Agilent-authorized field service engineer.”
Operation
Both the RF supply circuit breaker and the instrument power switch must be switched on for the instrument to work.
Operators must verify both power points during startup to avoid false no-power diagnosis.
Source
“Both the RF supply circuit breaker and instrument power switch must be switched on for the instrument to work.”
Operation
The yellow Plasma emergency off button is intended only for emergencies such as a melting torch, not for routine plasma extinction.
Routine use of the emergency off button requires manual reset and is not the correct shutdown method.
Source
“The yellow Plasma emergency off button, located next to the mains power switch on the front of the instrument, is designed to be used only in an emergency—for example if the torch is melting.”
Operation
Gas supplies should be left on at all times except during long periods of non-use to maintain uninterrupted purge of the polychromator assembly.
Continuous purge minimizes moisture ingress and preserves optical performance.
Source
“It is strongly recommended that gas supplies be left on at all times except during long periods of non-use.”