Unpacking, power, computer connection, startup and verification
Procedurepp. 17-34
Background collection, sample measurement and accessory cleaning procedures
Orderingpp. 35-36
Replacement parts and ordering information
Specpp. 37-39
Safety information and technical specifications
p. 40
Document conventions and navigation guide
p. 14table, 6 rows
p. 7table, 3 rows
p. 35table, 3 rows
p. 12table, 8 rows
Table 2 Diagnostic values
Value
Optimal (green)
Marginal (yellow)
Critical (red)
Comment
Energy (centerburst)
26,000–21,000
>30,000or < 15,000
> 31,000 or < 10,000
Indicates the overall alignment and proper gain adjustment for the system.
Battery life
AC or >30 min
30–5 min
<5 min
Source
1.9 A
>2.2 or <1.6
>2.5 or <1.0
Indicates a problem with the source control voltage or that the source has burnt out. Both voltage and amperage are displayed; however, amperage provides a sufficient diagnostic.
Laser signal
15,000–3,000
>15,000 or <3,000
>17,000 or < 2,000
Can verify gross alignment errors even if the reflectance cap is not in place.
Detector temperature
35–44
<35 or >44
<30 or >48
Indicates a problem with the cooling circuitry or an ambient temperature over the specified range.
CPU temperature
10–75
>75
>80
Indicates the ambient temperature over the specified range.
Agilent 5500t FTIR packing list
Item
One Agilent 5500t FTIR mid-IR spectrometer system, including TumblIR sampling accessory
One Agilent 5500 Series FTIR Operation Manual
One power cord
Spare Parts
Part Number
Description
430 - 0001
Agilent 5500 Series FTIR power supply
430 - 0018
Agilent 5500 Series FTIR USB cable
0020 - 900
Surfactant kit
For sampling accessories, service contracts, repair and refurbishment services, contact Agilent.
Table 1 Power switch LED status
Status
LED color
Action (% duty cycle)
System OFF
No LED illumination
N/A
System OFF
Red
100%
System startup
Red/green
Green 0.5 s on / Red 0.5 s on
System ON
Red/green
100%
Low battery
Green
50% (0.5 s on / 0.5 s off)
Critically low battery
Red
50% (0.5 s on / 0.5 s off)
Dead battery
No LED illumination
N/A
Firmware update
Red/green
Green blinks rapidly twice then red blinks rapidly twice
Engineering Facts
14 facts · every claim linked to a page
Installation
The instrument must be placed on a flat, stable surface away from hot surfaces and electromagnetic interference sources.
Prevents measurement instability and damage during installation.
Source
“The instrument must be kept away from hot surfaces and any sources of electromagnetic interference.”
Installation
Connect the power cord to the spectrometer first, then to the AC outlet to avoid startup damage.
Wrong connection order can cause power startup damage.
Source
“connect the power cord to the spectrometer system first and then to the power outlet.”
Installation
A UL-approved surge protection strip must be used between the outlet and the system power cord.
Protects the instrument from power surges and faulty power sources.
Source
“always use a UL approved power surge protection strip between the power outlet source and the system power cord.”
Installation
The 5500t FTIR must connect directly to a computer USB port, not through a hub or adapter.
USB hubs or adapters may cause unreliable communication with the instrument.
Source
“The 5500t FTIR must be directly connected to the computer USB slot and NOT through a USB hub or adapter.”
Operation
The system requires a 5-minute warmup before analysis may begin.
Starting analysis too early may yield invalid spectra.
Source
“The system must have a warmup period of 5 minutes, before analysis may begin.”
Performance
Diagnostic values should be checked daily or weekly, and performance validation quarterly or bi-annually.
Defines the routine verification schedule needed to trust measurement results.
Source
“diagnostic values should be checked on a daily or weekly basis.”
Performance
The system should be warmed up for at least 30 minutes before running performance validation tests.
Insufficient warmup can cause validation tests to fail or give misleading results.
Source
“The 5500 Series FTIR system should be warmed up for at least 30 minutes before conducting any of these tests.”
Maintenance
ZnSe windows tolerate only samples with pH between 4 and 9; strong acids or bases damage them.
Prevents permanent damage to the sampling windows and avoids warranty issues.
Source
“Only measure samples with a pH between 4 and 9.”
Maintenance
The TumblIR must not be used with solid samples because solids damage the ZnSe windows.
Using solids can destroy the windows and alter the fixed pathlength.
Source
“the TumblIR is not to be used with any solid samples, such as hard graphite lubricant.”
Maintenance
Only soft cotton cloth or swabs with acetone should be used to clean ZnSe windows; Kimwipes are abrasive.
Abrasive materials scratch the ZnSe windows and degrade optical performance.
Source
“Materials such as cotton swabs dipped in acetone are recommended for cleaning.”
Selection
The single reflection ATR suits high-absorbing samples and powders/solids because it has a sample press.
Choosing the wrong ATR variant gives poor spectra or cannot press solid samples.
Source
“The single reflection ATR is most suitable for higher absorbing samples such as rubber, polymers, paints and fibers.”
Selection
The triple reflection ATR is for low-absorbing liquids and pastes only and has no sample press.
Using solids on the triple reflection ATR gives bad contact and invalid results.
Source
“Only liquid or paste samples should be used with the triple reflection ATR accessory.”
Operation
The ATR sample press applies 15 pounds of force assuming a 0.03 inch sample thickness.
Knowing the press force helps ensure reproducible contact for powder and solid samples.
Source
“The spring-loaded sample press tip is set to apply 15 pounds of force assuming a 0.03 inch sample thickness.”
Operation
The DialPath background must be measured at the same pathlength as the sample.
Mismatched pathlengths produce invalid absorbance results.
Source
“The background measurement should be taken at the same pathlength that the sample will be measured at.”