Introduces trace permanent gas analysis and document purpose
p. 3
Presents experimental results and discussion of findings
Specpp. 4-5
Repeatability and detection limits in dual-column mode
Specp. 6
Analytical repeatability under low ionization energy conditions
Specp. 7
Linearity performance in dual-column mode
Configp. 8
Operation and results in single-column mode
p. 9
Summarizes findings and key takeaways
p. 10
Company website and contact information
p. 4table, 8 rows
p. 2table, 11 rows
p. 3table, 8 rows
p. 6table, 9 rows
Table 4. Repeatability and detection limits of selected target gases.
Sample name
CO2 Peak Area
CO2 RSD (%)
CO2 MDL (ppb, v/v)
N2O Peak Area
N2O RSD (%)
N2O MDL (ppb, v/v)
C2H4 Peak Area
C2H4 RSD (%)
C2H4 MDL (ppb, v/v)
SF6 Peak Area
SF6 RSD (%)
STD 25 ppb-1
838
2.5
2.2
140
2.5
2.0
128
3.3
2.1
3105
0.7
0.63
29
1.5
1.2
395
4.9
4.1
STD 25 ppb-2
842
136
123
3091
29
383
STD 25 ppb-3
853
139
125
3127
29
391
STD 25 ppb-4
830
133
120
3096
29
352
STD 25 ppb-5
815
135
118
3067
29
386
STD 25 ppb-6
802
138
119
3080
28
345
STD 25 ppb-7
802
130
117
3053
28
385
STD 25 ppb-8
803
139
119
3094
28
378
RSD (%) и MDL (ppb, v/v) указаны только для первой инъекции каждого газа; в остальных строках эти ячейки пустые, как напечатано.
Table 1. Composition of the standard gas mixture used in the experiment.
Number
Name
Concentration (ppm, v/v)
Gas Type
1
Hydrogen (H2)
4.95
Permanent
2
Oxygen (O2)
5.02
Permanent
3
Nitrogen (N2)
5.03
Permanent
4
Argon (Ar)
4.97
Permanent
5
Methane (CH4)
4.96
Permanent/greenhouse
6
Carbon monoxide (CO)
5.00
Permanent
7
Carbon dioxide (CO2)
5.01
Greenhouse
8
Nitrous oxide (N2O)
5.04
Greenhouse
9
Ethylene (C2H4)
4.96
Common chemical
10
Sulfur hexafluoride (SF6)
4.97
Greenhouse
11
Helium (He)
Balance gas
Permanent
Table 2. GC conditions.
Parameter
Value
Instrument
Agilent 8890 GC
Valve Oven Temperature
50 °C
Inlet Temperature
100 °C
Carrier Gas
High-purity helium (purity 99.9995%)
Analytical Column
Column 1: Agilent J&W Pora-BOND Q, 25 m × 0.25 mm × 3 µm (p/n CP7348PT)
Column 2: Agilent J&W CP-Molsieve 5 Å, 25 m × 0.25 mm × 30 µm (p/n CP7533)
Column Flow
Constant flow, 1.2 mL/min
Column Oven Ramp Program
Dual-column mode: 40 °C for 4.5 min, then 50 °C/min to 110 °C and 110 °C for 6 min
Single-column mode: 35 °C for 10 min
Transfer Line Temperature
230 °C
Note: Both are PLOT columns with integrated particle traps
Table 5. Repeatability data for 10 gases at low ionization energy.
Target Gas
CO2
N2O
C2H4
SF6
H2
Ar
O2
N2
CH4
CO
Number
Peak Area RSD (%)
Peak Area RSD (%)
Peak Area RSD (%)
Peak Area RSD (%)
Peak Area RSD (%)
Peak Area RSD (%)
Peak Area RSD (%)
Peak Area RSD (%)
Peak Area RSD (%)
Peak Area RSD (%)
1
1958 1.8
3141 2
1059 2
389 5.1
45 4.7
808 2.3
1728 2.9
384 4.8
1314 2.1
1314 1.7
2
1898
3060
1049
385
45
796
1756
390
1296
1296
3
1880
3029
1030
376
44
785
1644
354
1284
1284
4
1893
3026
1033
376
45
787
1727
390
1291
1291
5
1918
2996
1036
363
44
788
1644
358
1298
1298
6
1976
2968
1048
333
48
828
1726
395
1309
1309
7
1903
2974
1002
369
49
764
1638
360
1247
1247
8
1938
2961
1003
349
48
786
1660
355
1273
1273
Engineering Facts
14 facts · every claim linked to a page
Selection
The method requires a 5977C GC/MSD with high-efficiency source (HES) and dual-column valve-switching configuration to cover all 10 target gases.
Defines the minimum instrument configuration needed to reproduce the dual-column method.
Source
“The Agilent 8890 GC and 5977C GC/MSD equipped with HES and dual columns (as shown in the schematic in Figure 1) were used.”
Installation
Carrier gas must be high-purity helium of 99.9995% or better, with a helium purifier in the flow path.
Gas purity and purification directly affect detection limits and column lifetime.
Source
“Carrier Gas High-purity helium (purity 99.9995%)”
Selection
The dual-column setup uses a PoraBOND Q PLOT column plus a CP-Molsieve 5 Å PLOT column, both with integrated particle traps.
Column choice determines which gases separate chromatographically versus by MS only.
Source
“Both are PLOT columns with integrated particle traps”
Operation
H2 must be detected at low ionization energy to avoid helium carrier interference, while the other nine gases are best run at 70 eV.
Ionization energy selection is critical: wrong setting makes H2 undetectable or reduces sensitivity.
Source
“70 eV (low ionization energy for H2 detection to minimize interference from He)”
Operation
In dual-column mode CO and N2 are fully separated so m/z 28 can quantify both, but O2 and Ar coelute and must be distinguished by their fragment ions.
Tells the operator which gases need MS deconvolution rather than chromatographic resolution.
Source
“Although O2 and Ar coeluted, their distinct fragment ions enabled unambiguous differentiation based on the MS results.”
Performance
At 25 ppb the six quantified gases gave peak area RSDs of 0.7–4.9% with MDLs all below 4.1 ppb.
Establishes the achievable repeatability and detection limits for method validation.
Source
“the peak area RSDs (%) of six target gases, including CO2, N2O, C2H4, SF6, CH4, and CO, ranged from 0.7 to 4.9%, and their corresponding MDLs were all below 4.1 ppb”
Installation
Reliable ppb-level quantification of O2, N2 and Ar requires a purged valve oven to prevent air contamination during valve switching.
Hardware addition needed if trace O2/N2/Ar quantification is a requirement.
Source
“For precise quantification of these three gases at ppb levels, a purged valve oven is recommended to eliminate air contamination during valve switching.”
Performance
Under low ionization energy all 10 gases showed peak area RSDs of 1.7–5.1% at 5 ppm.
Confirms reproducibility of the low-energy mode used for H2 analysis.
Source
“the peak area RSD (%) for all 10 target gases ranged from 1.7 to 5.1%, demonstrating excellent reproducibility”
Performance
H2 detection limit under low ionization energy is about 500 ppb with 6% peak area RSD at that level.
Sets the realistic lower bound for H2 measurement and whether the method suits the application.
Source
“the S/N ratio of H2 was close to 3, indicating an MDL of approximately 500 ppb for H2 under low ionization energy conditions”
Performance
Six gases are linear from 10 to 5000 ppb with R2 above 0.9993 using forced-through-origin calibration.
Defines the validated calibration range and curve type for these analytes.
Source
“the linear correlation coefficients (R 2) for all six target gases were greater than 0.9993 across the 10 to 5000 ppb concentration range”
Operation
Single-column mode finishes in 7.5 minutes but coelutes H2, O2, Ar, N2 and CO, forcing use of low-abundance ions for N2 and CO.
Explains the speed/sensitivity trade-off when choosing single- versus dual-column mode.
Source
“in this mode, five target gases, including H2, O2, Ar, N2, and CO, coeluted and had to be identified by their distinct fragment ions”
Performance
CO detection limit in single-column mode is about 50 ppb, notably worse than in dual-column mode.
Quantifies the sensitivity penalty of single-column mode for CO.
Source
“it was inferred that the method detection limit of CO was approximately 50 ppb (v/v)”
Selection
Single-column mode is sufficient if H2 is not required and CO in the sample exceeds 50 ppb.
Direct guidance for choosing the simpler, faster configuration.
Source
“if H2 detection is not required and the target CO concentration in the sample is above 50 ppb, dual-column mode is not necessary, and single-column mode is sufficient to meet analytical requirements”
Installation
An optional backflush module can be added to remove H2O and heavy compounds from the system.
Informs hardware selection for samples containing water or heavy matrix.
Source
“an optional backflush module can be integrated to eliminate H2O and heavy compounds”