Analysis of Trace Permanent Gases

Contents

  1. Overview pp. 1-2
    Introduces trace permanent gas analysis and document purpose
  2. p. 3
    Presents experimental results and discussion of findings
  3. Spec pp. 4-5
    Repeatability and detection limits in dual-column mode
  4. Spec p. 6
    Analytical repeatability under low ionization energy conditions
  5. Spec p. 7
    Linearity performance in dual-column mode
  6. Config p. 8
    Operation and results in single-column mode
  7. p. 9
    Summarizes findings and key takeaways
  8. p. 10
    Company website and contact information
  9. p. 4 table, 8 rows
  10. p. 2 table, 11 rows
  11. p. 3 table, 8 rows
  12. p. 6 table, 9 rows

Table 4. Repeatability and detection limits of selected target gases.

Sample nameCO2 Peak AreaCO2 RSD (%)CO2 MDL (ppb, v/v)N2O Peak AreaN2O RSD (%)N2O MDL (ppb, v/v)C2H4 Peak AreaC2H4 RSD (%)C2H4 MDL (ppb, v/v)SF6 Peak AreaSF6 RSD (%)
STD 25 ppb-18382.52.21402.52.01283.32.131050.70.63291.51.23954.94.1
STD 25 ppb-2842136123309129383
STD 25 ppb-3853139125312729391
STD 25 ppb-4830133120309629352
STD 25 ppb-5815135118306729386
STD 25 ppb-6802138119308028345
STD 25 ppb-7802130117305328385
STD 25 ppb-8803139119309428378

RSD (%) и MDL (ppb, v/v) указаны только для первой инъекции каждого газа; в остальных строках эти ячейки пустые, как напечатано.

Table 1. Composition of the standard gas mixture used in the experiment.

NumberNameConcentration (ppm, v/v)Gas Type
1Hydrogen (H2)4.95Permanent
2Oxygen (O2)5.02Permanent
3Nitrogen (N2)5.03Permanent
4Argon (Ar)4.97Permanent
5Methane (CH4)4.96Permanent/greenhouse
6Carbon monoxide (CO)5.00Permanent
7Carbon dioxide (CO2)5.01Greenhouse
8Nitrous oxide (N2O)5.04Greenhouse
9Ethylene (C2H4)4.96Common chemical
10Sulfur hexafluoride (SF6)4.97Greenhouse
11Helium (He)Balance gasPermanent

Table 2. GC conditions.

ParameterValue
InstrumentAgilent 8890 GC
Valve Oven Temperature50 °C
Inlet Temperature100 °C
Carrier GasHigh-purity helium (purity 99.9995%)
Analytical ColumnColumn 1: Agilent J&W Pora-BOND Q, 25 m × 0.25 mm × 3 µm (p/n CP7348PT) Column 2: Agilent J&W CP-Molsieve 5 Å, 25 m × 0.25 mm × 30 µm (p/n CP7533)
Column FlowConstant flow, 1.2 mL/min
Column Oven Ramp ProgramDual-column mode: 40 °C for 4.5 min, then 50 °C/min to 110 °C and 110 °C for 6 min Single-column mode: 35 °C for 10 min
Transfer Line Temperature230 °C

Note: Both are PLOT columns with integrated particle traps

Table 5. Repeatability data for 10 gases at low ionization energy.

Target GasCO2N2OC2H4SF6H2ArO2N2CH4CO
NumberPeak Area RSD (%)Peak Area RSD (%)Peak Area RSD (%)Peak Area RSD (%)Peak Area RSD (%)Peak Area RSD (%)Peak Area RSD (%)Peak Area RSD (%)Peak Area RSD (%)Peak Area RSD (%)
11958 1.83141 21059 2389 5.145 4.7808 2.31728 2.9384 4.81314 2.11314 1.7
218983060104938545796175639012961296
318803029103037644785164435412841284
418933026103337645787172739012911291
519182996103636344788164435812981298
619762968104833348828172639513091309
719032974100236949764163836012471247
819382961100334948786166035512731273

Engineering Facts

14 facts · every claim linked to a page
Selection

The method requires a 5977C GC/MSD with high-efficiency source (HES) and dual-column valve-switching configuration to cover all 10 target gases.

Defines the minimum instrument configuration needed to reproduce the dual-column method.

Source
“The Agilent 8890 GC and 5977C GC/MSD equipped with HES and dual columns (as shown in the schematic in Figure 1) were used.”
Installation

Carrier gas must be high-purity helium of 99.9995% or better, with a helium purifier in the flow path.

Gas purity and purification directly affect detection limits and column lifetime.

Source
“Carrier Gas High-purity helium (purity 99.9995%)”
Selection

The dual-column setup uses a PoraBOND Q PLOT column plus a CP-Molsieve 5 Å PLOT column, both with integrated particle traps.

Column choice determines which gases separate chromatographically versus by MS only.

Source
“Both are PLOT columns with integrated particle traps”
Operation

H2 must be detected at low ionization energy to avoid helium carrier interference, while the other nine gases are best run at 70 eV.

Ionization energy selection is critical: wrong setting makes H2 undetectable or reduces sensitivity.

Source
“70 eV (low ionization energy for H2 detection to minimize interference from He)”
Operation

In dual-column mode CO and N2 are fully separated so m/z 28 can quantify both, but O2 and Ar coelute and must be distinguished by their fragment ions.

Tells the operator which gases need MS deconvolution rather than chromatographic resolution.

Source
“Although O2 and Ar coeluted, their distinct fragment ions enabled unambiguous differentiation based on the MS results.”
Performance

At 25 ppb the six quantified gases gave peak area RSDs of 0.7–4.9% with MDLs all below 4.1 ppb.

Establishes the achievable repeatability and detection limits for method validation.

Source
“the peak area RSDs (%) of six target gases, including CO2, N2O, C2H4, SF6, CH4, and CO, ranged from 0.7 to 4.9%, and their corresponding MDLs were all below 4.1 ppb”
Installation

Reliable ppb-level quantification of O2, N2 and Ar requires a purged valve oven to prevent air contamination during valve switching.

Hardware addition needed if trace O2/N2/Ar quantification is a requirement.

Source
“For precise quantification of these three gases at ppb levels, a purged valve oven is recommended to eliminate air contamination during valve switching.”
Performance

Under low ionization energy all 10 gases showed peak area RSDs of 1.7–5.1% at 5 ppm.

Confirms reproducibility of the low-energy mode used for H2 analysis.

Source
“the peak area RSD (%) for all 10 target gases ranged from 1.7 to 5.1%, demonstrating excellent reproducibility”
Performance

H2 detection limit under low ionization energy is about 500 ppb with 6% peak area RSD at that level.

Sets the realistic lower bound for H2 measurement and whether the method suits the application.

Source
“the S/N ratio of H2 was close to 3, indicating an MDL of approximately 500 ppb for H2 under low ionization energy conditions”
Performance

Six gases are linear from 10 to 5000 ppb with R2 above 0.9993 using forced-through-origin calibration.

Defines the validated calibration range and curve type for these analytes.

Source
“the linear correlation coefficients (R 2) for all six target gases were greater than 0.9993 across the 10 to 5000 ppb concentration range”
Operation

Single-column mode finishes in 7.5 minutes but coelutes H2, O2, Ar, N2 and CO, forcing use of low-abundance ions for N2 and CO.

Explains the speed/sensitivity trade-off when choosing single- versus dual-column mode.

Source
“in this mode, five target gases, including H2, O2, Ar, N2, and CO, coeluted and had to be identified by their distinct fragment ions”
Performance

CO detection limit in single-column mode is about 50 ppb, notably worse than in dual-column mode.

Quantifies the sensitivity penalty of single-column mode for CO.

Source
“it was inferred that the method detection limit of CO was approximately 50 ppb (v/v)”
Selection

Single-column mode is sufficient if H2 is not required and CO in the sample exceeds 50 ppb.

Direct guidance for choosing the simpler, faster configuration.

Source
“if H2 detection is not required and the target CO concentration in the sample is above 50 ppb, dual-column mode is not necessary, and single-column mode is sufficient to meet analytical requirements”
Installation

An optional backflush module can be added to remove H2O and heavy compounds from the system.

Informs hardware selection for samples containing water or heavy matrix.

Source
“an optional backflush module can be integrated to eliminate H2O and heavy compounds”